Hi,
I am working on a Fluke 8840A/AF meter which has a rather puzzling problem when using the rear inputs.
A few details:
1. When using the front inputs, everything works just fine.
2. If I switch to the rear inputs, DCV readings are out of whack and nothing else really works - e.g. when I press 2W resistance or 4W resistance, I get error 31 (which, according to the manual, should be displayed if I tried to enable current measurements, either DC or AC, with the rear inputs enabled).

Things I've checked:
a. Power rails - all check good, either with front inputs selected, or with the rear ones.
b. F/R switch - checks good.
c. F/R sense signal to the CPU (i.e. U202) - checks good (i.e. pulled low when rear inputs are selected).
d. CPU - I've swapped the CPU with another one (as well as the associated U222 ROM IC), but the problem remains.
e. Swapped U212 with the same IC from a donor board (no change).
f. Swapped U219 with the same IC from a donor board (same situation).
g. Socketed ICs - re-seated all socketed ICs, with no difference.
h. Intermittent contacts - poked the board and a few components, but no change.
i. Checked U222 contents and they are good.
The only lead I have right now is the strange behavior when pressing 2W or 4W resistance buttons - i.e. getting error 31.
That would make me think there must be some logic fault somewhere, so the CPU interprets the 2W and 4W button presses as mA DC presses.
But, then, why does this happen only when the rear inputs are selected?
What else changes when the F/R switch is set for the rear inputs (other than the path of the input signal)?
Any suggestions are appreciated.
Thanks!