Hi squadchannel
Thank you very much for sending the files. I have successfully extracted them and the instrument now recognizes the hard drive and the software is running.
However, the unit is showing the error “Error: Self gain cal (0.00 dB)”. I tried enabling the CAL signal in the Input menu — the 25 MHz signal appears correctly, but the 100 MHz signal does not show up.
I also tested external signals: frequencies below 40 MHz are OK, but from 40 MHz to 50 MHz the instrument reports A/D OVERFLOW and the signal behavior is very abnormal. Above 50 MHz up to 8 GHz, the unit does not detect any signal at all.
It seems that I still need to resolve some hardware issues before the analyzer can operate normally again.
Anyway, thank you very much for your help with the software side.