I know with certainty that the calibration values are stored in the Dallas NVRAM because dead batteries is the reason the scopes I have need calibration.
Hi James ,
First you don't need to spend $20K unless you feel like giving me a BD present.
.
If you Have lost the CaL . The battery must be replace if not already done so. as they some times
die short circuit
If you managed to do all the voltage calibrations OK . Though the 50ohm terminator . you guessed that
when it ask for 1v you have to set the function gen to 2v so the scope see 1v @ 50
(Example)
When it ask for the fast rise pulse this also must be doubled so the scope see 1v pulse @ 50
. (example)
If using a UTG 962 set to 50ohm load the out put voltage is doubled .corrected for 50
put a bench meter on the function gen on the DC 1.0000v and measured what the voltage error is
from the function Gen (example) set dc level to 1.0000v Meter read 1.0010 . so minus the error on the
function gen . So when set to 2.0000 volt - 0.0020 to correct . (use a RG cable with BNC to Banana )
Don't measure with single jumper wires .
then set function gen to pulse and the corrected voltage and all should be good.
The TDS has built in ref voltages & frequencies these have to be matched before it will load them back
into NVRam again .
If this still fails then you may have a few out of tolerance caps . See photo above the Blue caps were
out on my one. these were caps I replaced 1% mylar Good enough and this solved that .
I guess you don't need NASA spec's . After Cal I checked with a Leo Bodnar GPSDO .. less than 1% error.
Or you can raise or lower the signal voltage by just a few uv will do it as well to compensate for wear and tear.
The one Most important thing that I found that this scope is very fussy on noisy connections during Cal setup
the connections MUST Be 100% good connections including the grounding.
As any voltage drop on the cable will result as a fail
Did it to me as well .
The Battery MUST be replaced or you will be recaling every other day .
or never unplug it from the mains .
If you use the UTG962 Set to 50ohm Load the voltage will be correct Set@987mv =1.9990v
Scope sees after terminator 1.0005v Adjust to match cable losses (updated typo)
Did this Help or did I miss understand something with the Question