Electronics > Repair
[FIXED] Venerable AIM TTI PL303QMD-P needs help - Overvoltage trip error
Zucca:
Okay, the problem is in one the digital PIC board (originally master).
Swapped the two and the error follows the PIC board, the other side was fixed.
Two corrections from before:
1) the misaligned molex (on the master power board) does not play a role... it is correct in both position misaligned or not..
2) the two power boards are identical but there is a hard jumper in the master (LK1)... so for basic tests you can swap them temporarily but them put them back in place.
getting closer!
Zucca:
Sorry for the delay, I am selling some stuff (go to the buy/sell section to check it out) and my 2015 Ford F150 needed some immediate attention.
Will resume this adventure soon, I am confident I will fix it.
Zucca:
uP board has some digital and analog stuff.
As a quick check the analog parts looks good.
Next I will check the DAC and ADC with a scope.
Can't be the uP because the display "Output trip" is sent by the uP iself.
Gertjan:
Hi Zucca, good to hear that you are still at it! :-+
This very nice power supply needs to be in use on your bench, instead of sitting in your repair que :)
m k:
On the bench already.
MCU can have separated on/off info pins for both channels.