Electronics > Repair
[FIXED] Venerable AIM TTI PL303QMD-P needs help - Overvoltage trip error
m k:
Dang, too late.
My guess was that OUP is too wide as output.
In case you have enough courage you can try creating an OCP trip using the functioning side of the machine.
The feedback can be on/off style controller pins or analog inputs and then controller decides what is bad.
Older version has ILIMIT as a single pin, but can't say what it actually is.
Triple model 5V has OVP trip signal, others have OPON.
Swapping boards and using brute force measuring tactics you can "easily" narrow possible problem pins.
But schematics would be better.
For PJ6 my guess is that it doesn't help.
But who knows, maybe it's constantly outputting some console info.
Zucca:
Thank m k to stay with me in this madness.
After some emails, TTi support told me they believe/think the uP must be defective and that's it, they did not respond to my last email.... I let them go in peace...
They helped me a lot for free!
Despite that I think we have still cards to play, now that we know that it's an OVP problem we can dig into the right direction.
Please be aware, if you see a OVP malfunction in your device, I would say 90% the bad guy is the power board, my case is an unicorn IMHO.
According to my rev eng, there are 2 places where the output voltage (aka vsense+ and vsense-) comes into the control board:
1 - into the ADC
2 - into the VC generation analog part, basically to close the output voltage regulation feedback loop:
Vset ----> Vc --> Output ----
| |
--------------------------
- Vc is the control voltage
- Vset is the target voltage coming from the uP
Please note the Vset to Vc analog part is on the control board, the Vc to Output is on the power board.
Now, two facts:
a) I can't find any problem in the analog control board [CB] part
b) after powering up the unit, immediately the OVP is triggered on the side (master or slave does not matter) where the offending control board is installed.
Here below the startup sequence on the offending side
Green is the 5V rail
Blue is Vc
Yellow is Vset
this means to me:
1) Assuming the uP is still alive since I see "OUP tRIP" in the display, we need to check what it is going on between Vsenses lines before and after the ADC.
2) I need to compare the same startup sequence on the good side with output off, to see if Vc going to the roof for about 85ms after startup is normal.
I think this is not a problem... Vset stay at zero.... so...
[oh I did not check if the output is spiking at startup as well... but since Vset is 0 all the time, I do not expect any movement there.]
Before the ADC this is what is happening:
where A17 and A16 are the two ADC inputs, simulated in QSpice we get:
it's too late now, I need to sleep.... more to come.
m k:
Chopped startup of PID controller.
First D goes through the roof, then 'I' start kicking in and finally everything is steady again.
Maybe this startup shouldn't go through the feedback.
Zucca:
Back to the ADC section, this is the correct output sense circuit. The previous one is not good.
basically the Vsense+ is connected to GND (when sense is set to local of course).
And this is the output voltages to the ADC (sweep 0V to 35V output voltage):
And this make sense, because the ADC input range is 0V-2.5V, in my previous one input was above 2.5V so it would had not work well.
In my previous measurements, I confirmed those ADC inputs lines are sitting at 2.4V when the outputs are disabled, which again matches the simulation.
Zucca:
Startup voltages, yellow output, green 5V rail
since the output doesn't move, again either the ADC or the uP (unlikely) are bad.
Next step sniff the ADC SPI
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version