I just got my HM605 out of storage, and I realize it has some problems measuring vertical tension : when checking 5V DC rail of a switching power supply the trace barely gets to 4.1V.
Same thing happens on both channels.
All buttons are depressed or set in calibration mode.
I checked voltages on the Y board (expected -> measured) :
+68V -> +68V
+4.65 -> +4.1V
+140V -> +139.1V
+4.3V -> -7.5V
+12V -> +12V
+5V -> +5V
-12V -> -11.9V
The only weird reading is the +4.3V one, but the manual specifies "-12V antipole" in that case, and I don't know what it's about and if it has anything to do with my problem ...
I also checked the impact of VR501 on the YF board (ie Y final amplifier gain), but it only variates of 0.2V max, which fall far short from the mark.
Do you have an idea where I should look, a direction to point to ? I'm not an oscilloscope expert, but this problem is bothering me and I would love to be able to fix it.