Author Topic: Vertical Test Group Self Test Failed on Agilent MSO8104A  (Read 1696 times)
Offline admiralmaggieTopic starter
« on: April 27, 2017, 05:41:46 pm »
I'm getting "Vertical Test Group" Failed on my Agilent MSO8104A. It does not provide any other information at all. The service manual says it should specify exactly which test under Vertical Test Group failed but the error log doesn't say anything other than "Vertical Test Group: Failed". All other tests are passing and scope appears to be functioning properly. Any thoughts on this?
Re: Vertical Test Group Self Test Failed on Agilent MSO8104A
« Reply #1 on: April 27, 2017, 07:43:56 pm »
Here is more details. I found the log file. All values look nominal... Any help is greatly appreciated:
** Test Run **   Scope Tests --- 27 APR 2017 13:12:01
         Reference setting for gain ratio
         Preamp Settings Chan 1 , CPL=DC, IMP=HiZ, BWL=OFF
         ****INPUT=G1, Mux=X1, GainSet1=X1.6, GainSet2=X2.6, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.614407&E, Expected Ratio(+/-20%)=2.600000
         ****INPUT=G1, Mux=X1, GainSet1=X1.6, GainSet2=X1, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.009772&E, Expected Ratio(+/-20%)=2.000000
         ****INPUT=G1, Mux=X1, GainSet1=X1.6, GainSet2=X2.6, Vernier=255
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.615183&E, Expected Ratio(+/-20%)=1.600000
         ****INPUT=G1, Mux=X1, GainSet1=X1, GainSet2=X2.6, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.029605&E, Expected Ratio(+/-20%)=2.000000
         ****INPUT=G1, Mux=X2, GainSet1=X1.6, GainSet2=X2.6, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B0.040111&E, Expected Ratio(+/-20%)=0.040300
         ****INPUT=G2, Mux=X1, GainSet1=X1.6, GainSet2=X2.6, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.607465&E, Expected Ratio(+/-20%)=2.600000
         ****INPUT=G2, Mux=X1, GainSet1=X1.6, GainSet2=X1, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.613885&E, Expected Ratio(+/-20%)=1.600000
         ****INPUT=G2, Mux=X1, GainSet1=X1, GainSet2=X2.6, Vernier=0
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.026905&E, Expected Ratio(+/-20%)=2.000000
         ****INPUT=G2, Mux=X2, GainSet1=X1.6, GainSet2=X2.6, Vernier=0
         Preamp Settings Chan 2 , CPL=DC, IMP=HiZ, BWL=OFF
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.599576&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.011475&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.608126&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.028099&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B0.039720&E, Expected Ratio(+/-20%)=0.040300
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.601717&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.607186&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.024429&E, Expected Ratio(+/-20%)=2.000000
         Preamp Settings Chan 3 , CPL=DC, IMP=HiZ, BWL=OFF
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.610994&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.024590&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.612272&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.031250&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B0.040159&E, Expected Ratio(+/-20%)=0.040300
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.607963&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.616540&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.026127&E, Expected Ratio(+/-20%)=2.000000
         Preamp Settings Chan 4 , CPL=DC, IMP=HiZ, BWL=OFF
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.624190&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.025000&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.615691&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.038591&E, Expected Ratio(+/-20%)=2.000000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B0.039324&E, Expected Ratio(+/-20%)=0.040300
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.603189&E, Expected Ratio(+/-20%)=2.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B1.611123&E, Expected Ratio(+/-20%)=1.600000
         Status: &C0&&B -- PASSED&E, Gain Ratio=&C4&&B2.019343&E, Expected Ratio(+/-20%)=2.000000
         Chan 1 Relay actuations:
         ***Imp Sel = 0 ***25x Atten = 0 ***10x Atten = 0
         Chan 2 Relay actuations:
         Chan 3 Relay actuations:
         Chan 4 Relay actuations:
         Desired Preamp Differential Output Voltage = 0.200V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 1 ) = 0.403V   (-Chan 1 ) = 0.604V
         Desired Preamp Differential Output Voltage = -0.200V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 1 ) = 0.604V   (-Chan 1 ) = 0.404V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 2 ) = 0.402V   (-Chan 2 ) = 0.602V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 2 ) = 0.603V   (-Chan 2 ) = 0.403V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 3 ) = 0.403V   (-Chan 3 ) = 0.603V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 3 ) = 0.604V   (-Chan 3 ) = 0.404V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 4 ) = 0.403V   (-Chan 4 ) = 0.603V
         Offset Dac = 0x0  Offset Range = 1     (+Chan 4 ) = 0.604V   (-Chan 4 ) = 0.404V
   Test: &B&UVertical Test Group&E&C13&&B -- FAILED&E
   Test: &B&UTrigger Test Group&E&B -- PASSED&E
         Chan 1 Slice Data = 88,94,90,97,91,89,87,89,89,88,96,88,91,94,92,96,
         Chan 2 Slice Data = 67,68,65,67,65,64,65,69,63,63,73,70,69,71,65,62,
         Slices in error = 0,1,2,3,4,5,6,7,8,9,10,11,12,13,14,15,16,17,18,19,20,21,22,23,24,25,26,27,28,29,30,31,
         Chan 3 Slice Data = 63,68,60,62,70,63,65,69,62,57,60,62,64,67,62,66,
         Chan 4 Slice Data = 64,72,66,64,63,71,70,64,64,66,69,66,61,63,63,66,
   Test: &B&UAdc Test Group&E&B -- PASSED&E
         Long Count = 105
         Long Count Residue = 148
         Short Count = 60
         Short Count Residue = 149
   Test: &B&UTimebase Test Group&E&B -- PASSED&E
   Test: &B&UAcq Interface Test Group&E&B -- PASSED&E
   Test: &B&UAcq Memory Group&E&B -- PASSED&E
         System Temp 36.7  Degrees C
   Test: &B&UMisc Scope Test Group&E&B -- PASSED&E
&BFinal Result: &E&B&UScope Tests&E&C13&&B -- FAILED&E
« Reply #2 on: May 04, 2017, 04:39:58 pm »
UPDATE: I ran Self Calibration a few times and it passes but the pesky Vertical Group test still fails... Any suggestions? :-BROKE