Hi folks,
I have a Palec VTVM, model TV-M. This meter has been repaired in the distant past a number of times. Many of the carbon resistors had since wandered way out of range, and so I have replaced all the capacitors and resistors.
Now everything appears to be working just fine, with the exception of the AC ranges which are reading about 50% low. I have checked the rotary switches for resistance in the contacts, the valve bases, anything at all that might be a cause. I have even gone so far as to test and check that I have put all the components in their right places. It seems I have.
Now oddly enough, the meter seems to use the same probe (with isolating resistor) for both AC and DC measurements. I have made up a direct probe for AC use, but the meter needle barely flickers and then returns to zero if I use it. The isolated probe (1meg) gives me the 50% low reading.
Adjusting the trimmer VR2 to the minimum resistance brings the needle up a little closer to the correct value, but nowhere near enough.
Both the 12AU7 and the 12AL5 test as good on an emission type tester. (Palec ET-4a)
Attached is my attempt at a schematic diagram. I think it's complete and correct.
So, if you can see it, please let me know where I'm going off the rails.
Thanks,
Clay.