I have a Wavetek 288 Military version function generator that I've owned for several years that has the strangest problem that is not corrected by either auto or manual calibration. No Matter what the waveform the amplitude is exactly 2X what the display shows and the same goes for the DC-offset.
For example, if I set the waveform to DC and offset to 1V the generator(and multi-meter) will output exactly 2V. If I set a waveform - irregardless of shape to 1VPP I will get 2VPP. Changing from VPP to VP or others are consistently the same - exactly 2X what the display shows. So to summarize, no matter what the waverform or offset, the actual values shown on my scope & multi-meter show 2X what is on the Wavetek display

I've checked that the line-voltage pc board is installed correctly, manually calibrated and verified(and adjusted R208) to 7.071v per the manual and all calibration tests pass w/o issue. The autocalibrate feature also passes w/o issue.
I've lived with this for some time and thought it might have something to-do with the military version. The waveforms all look great, minimal distortion and really love this generator so hope someone out there has a clue regarding this very strange problem. Because the output is exactly 2X whats on the display I'm thinking there is a simple explanation for this...