Author Topic: Wavetek 9100 blown power supply  (Read 1732 times)
Offline 002911Topic starter
  • Country: mx
« on: August 02, 2023, 04:23:52 pm »
Good afternoon, we have a Wavetek 9100 unit that during the test, when it is turned on, damages some fuses of the source (F501, F502, F503, F504).
These fuses blow near the end of the test and the screen goes blank.
Does anyone have any idea why this is happening?
Thank you.
Offline Alex_Baker
Re: Wavetek 9100 blown power supply
« Reply #1 on: August 03, 2023, 03:47:15 pm »
Do you have an idea of what the self test is supposed to do, specifically the point where it fails? That may lend some clues to where the problem is. My first though is that the device is attempting to test a part of itself that has failed, overloading the power supply.
Offline Gangway
« Reply #2 on: August 06, 2023, 10:39:12 am »
Do you have the manuals? If not there is tech info online for service notes. We have two 9100E’s at work, never had hardware fault.
One unit working, but second one out of spec on certain modes, so will list unit for sale. Fluke no longer support these units.
Offline alm
« Reply #3 on: August 06, 2023, 10:55:58 am »
Don't Fluke offer adjustment for the Wavetek 9100? I know they even refuse to calibrate some former Wavetek units without adjustment, which is strange since all they need to do is measure a bunch of values. Maybe some third party cal labs can do it? Or is it so far out of spec that it would require repair?
Offline Gangway
« Reply #4 on: August 06, 2023, 11:26:00 am »
The 9100E is a software calibrated unit. Ours have to get calibrated by a NATA Lab as required for certification of NATA Traceable Reports.
Too much work on to take time & try and investigate. Fluke USA, UK, EU, all say same thing re support for this unit.
Have replaced this unit with a Transmile Multifunctin Calibrator.
Offline Arhigos
  • Country: gr
« Reply #5 on: August 09, 2023, 12:15:52 pm »
« Last Edit: August 09, 2023, 01:01:57 pm by Arhigos »
The following users thanked this post: doktor pyta, coromonadalix
Offline Louis_C
  • Virtual lobotomies are now REAL!
    • Hamvoip RF/IP UHF repeater system
« Reply #6 on: September 27, 2023, 01:23:44 pm »
Hello all. I have a Fluke/Wavetek 9100A with a high current issue at the fan power connection. On the schematic, sheet 5 of 8, I replaced U502/503, R527 and C512 in the fan power circuit. I still have a very high current condition at the fan outputs at E508/E509. Does anyone have an idea as to what else to look at here? My calibrator is down and is needed. Thank you, Lou.
I reject your reality and insert my own hyper-dimensional space. Made with real artificial ingredients.
Offline coromonadalix
« Reply #7 on: September 28, 2023, 10:28:13 am »
if you remove the fan  are the 15volt lines ok ??
or the fan is getting stuck in the bushing / sleeve,  simply change the fan ?
Offline Louis_C
« Reply #8 on: September 28, 2023, 12:48:16 pm »
IF, I connect the fan, it will burn out. I'm not sure where the condition is coming from. I suspect the rectifier so, I'm going to replace that as well. I just don't see where else this could be coming from. BTW, the original fan died, it is very (crunchy) when you spin it by hand. I suspect the fan failed and pulled high current and caused a failure farther down the circuit. I'm not sure.