I found myself a Wavetek 9100 today, it cannot boot. Upon powerup the instrument turns on with a blank white display and stays there. I managed to find the service manual and schematic on https://ko4bb.com/.
So far i have confirmed that PSU voltages are correct. So now I am having a look at the digital processor board. Will be probing the SRAM for activity later today. I also saw its using a Dallas 1287 RTC.
Is there be a possibility a dead battery in it cause the unit to not turn on?. Any wavetek 9100 user out there that can advice me on this?
« Last Edit: April 29, 2022, 01:27:55 am by CrazyTiger »
Offline ebclr
Re: Wavetek/Datron 9100 Multifunction Calibrator Unable to boot
« Reply #1 on: April 29, 2022, 02:25:00 am »
Where did you find service manual, on the link only operational manual, I would love to have the real service manual
  • Country: my
« Reply #2 on: April 29, 2022, 03:11:19 am »
Well, I wont say its a conventional service manual, not sure how to name it, but here is the link:
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=06_Misc_Test_Equipment/Datron/Datron_9100
Progress Update:
I have already unsoldered the Dallas chip and drill a hole to probe for the battery voltage, its dead.
Will solder an external battery shortly to see if it does something.
Offline Testtech
  • Country: my
« Reply #4 on: May 01, 2022, 10:55:06 am »
This manual is for 9100A, a Digital Tester. Maybe fluke got stuck with Datron's model number when they rebadged and  ended up getting stuck with 2 different instrument with similar model number?
Anyway, it was not the Dallas RTC, it was a long shot to begin with. Will continue my diagnose :)
The following users thanked this post: ebclr
Offline jsolomonsc@gmail.com
« Reply #5 on: September 22, 2022, 12:24:16 am »
Make sure that the dip switches 6, 7 and 8 in the back are set properly.