I bought a broken CombiScope PM3380B and I repaired it however Flash-ROM is full. Ten saves have been done in the past. In service manual is saying that:
"The Flash-ROM’s contain blocks of calibration data (of which the most recent block is valid) and the
operating software. After operation of softkey ’save calibr data’, the text ’XX CALIBRATION FIELDS
FREE’ is displayed. XX can be a figure between 10 ... 1 or ’NO’. In case of ’NO’, the Flash ROM must
be emptied and redundant blocks of calibration data must be removed. To have this done, send your
oscilloscope to the nearest Service Center."
There is no Service Center near to my location. Is somebody willing to help me to sort out the issue and free up Flash-ROM? I will really appreciate any help/advice.