An AD slope error is more likely the AD hybrid or its controller (a micro) and rather 'un-obtainium'. As alex. said your pretty much going to have to look into it yourself. Download the manual and read through the section 7-d-38 and flowchart d (unfortunately they use an signature analyzer to test, but it gives you an idea of whats going on!). Look at the circuit, U403 is the AD and you may find something amiss around here.
Check the DC supply voltages for correct values and no mains ripple etc (you have another multimeter or CRO?), at least to eliminate bad power supplies. Check the AD reference is good U461.
Is the PCB clean, has this meter been abused, since it was cheep.
Does the error message all that happens, no other functionality possible?.
Good luck