Hi All,
I bought a used HP 34401A with some faults. When performing the self tests, almost all errors were reported, from 608 to 622
First thing, I got the Service Manual, started to probe power rails: bingo! CR305 was broken (open), meaning no +15VDC. It was the only missing power rail.
I used a bench supply to feed the missing +15VDC rail.
Still, the meter was showing errors 612 and 618 (all related to the 500nA current source), and sometimes, the error 614 would appear randomly.
For errors 6112 and 618, this forum quickly drew my attention to U201, although I was a little skeptic since the other current sources were fine.
I got myself a LM353 as a replacement (the only CMS my retailer had in stock that could replace the AD706).
Instead of using my external bench power supply, I replaced the broken CR305 with a non CMS component, at least temporarily.
Today, the meter fully passes all its tests!
I compared measures with another 344401: they're pretty close from each other for each function (VDC, VAC, IDC, ohm 2W (I didn't test 4W) and freq/period.
But !
I'm still noticing a weird effect:
- on VDC function, with no test lead connected, the meter keeps displaying random figures, around 0.220V (sometimes much higher !) .
If I connected a test lead in the LO input, same behaviour
If I connect a test lead in the HI input, and that I touch the tip of the test lead, the displayed voltage drops to 0.3mV, but still very noisy.
- on VDC, shorting the HI to LO input gets me to 0.0003mV (AFAIR)
- on 2W ohm, in AUTO range, the meter sets on the highest range and displays some random but stable figures sometimes around 20Mohms, sometimes about 80Mohms.
Again, shorting the inputs or putting a known resistance in between HI and LO seems to provide an acceptable reading.
- on ADC, putting a test lead into the HI input (yes, the HI input, where you're supposed to measure VDC or ohms) gets the display to 0.221A
- shorting the LO to the current input gets the display to almost 0mA.
This behaviour affects both the rear and the front inputs, and playing around with the front/rear switch have no effect.
Looks like a small current leaks from the HI to LO input, disturbing the meter with the lowest current source, hence displaying erroneous value in the ohm function.
I also noticed that on VDC function, once I stop shorting the HI to LO, the displayed values quickly increase from 0V to .... I don't, it keeps increasing.
I gently taped the components to find a potential bad solder joint (more particularly resistors R104 to R111, which are prone to bad joints as far as I read from this forum). No luck so far.
Any ideas?
Thanks for your help.