The Current Adjustment is actually Current LIMIT Adjustment, so its adjust the limit of the current. You can imagine a water pipe with 2 water tap/valves. One valve is adjusted by the landlord to half Open, the other valve is adjusted by the Tenant. So when the tenant fully open the valve, only half of the flowrate can be achieved. But if tenant adjust to lower than half, the water flow rate can varies between 0 to half. Got the idea? So depending on the load you put, the current can be less then or equal to the current limit set but no more.
To test your power supply.
1.0 Turn Current Adjustment and Voltage Adjustment Knobs to FULLY ANTICLOCKWISE - 0 limit
2.0 Hook a Ampere Meter Connected to measure 10A max onto the +ve and -ve output
3.0 Turn On - Observe the Current read 0 ampere.
4.0 Then slowly rotate voltage knob so that 1V ~ 2V showed up.
5.0 Then slowly rotate the current, SLOWLY rotate the current knob up and observe the Ampere Meter reads the current and tally with the power supply readout.
6.0 If you can adjust up to say 1A or the current limit of your PS, then consider the PS is OK.
If any abnormalities, then turn-off the power supply immediately and inform us back.