My situation is :
- the 7061 is not calibrated at all.
It had probably some EEPROM problems, I'm waiting another programmer to check it, but
now it has a new eeprom so no cal.
- the 7075 had many problems, I changed reed contacts, it has a bad pot on 10V dc calibration,
also I changed the chopper leds. So it's calibration isn't very trustable.
343 situation is:
343 owner isn't a professional seller, he works in an electronics company, and they
used it for some projects some years ago. They bought it used, and from last pro calibration,
they adjusted it, don't know why, so the original calibration is "compromised" anyway.
Their 8846 was calibrated by a pro service in 2012.
To resume his 8846 is recently calibrated in a professional laboratory,
my instruments aren't calibrated at all, the 343 lost it's original calibration.
I should at least find a way to minimize the errors on this procedure.
He is too far to go in person to his lab .
Fabio.