VPC300 particle counter
« on: March 23, 2023, 01:24:46 pm »
I am wondering what is the unit of particle counts in VPC300. Are these numbers (counts per particle size) are measured in per cubic meter ?
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6384
  • Country: ca
  • Non-expert
Re: VPC300 particle counter
« Reply #1 on: March 23, 2023, 10:05:32 pm »
ug/m3 is the standard measurement unit yes.

But it looks like it depends on what mode you are in:
"Particle Count Modes Cumulative, Differential, and Concentration modes"

https://www.fondriest.com/pdf/extech_vpc300_manual.pdf
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline joniengr081Topic starter

  • Regular Contributor
  • *
  • Posts: 159
  • Country: no
Re: VPC300 particle counter
« Reply #2 on: March 24, 2023, 06:59:05 am »
ug/m3 look like a unit of density but the measurements should be in terms of count per m3 or count per cm3, maybe ft3. I am not getting how it related to density.
 

Offline joniengr081Topic starter

  • Regular Contributor
  • *
  • Posts: 159
  • Country: no
Re: VPC300 particle counter
« Reply #3 on: March 24, 2023, 08:46:09 am »
The user manual says "If a room is classified as an ISO clean room or clean zone, the counts of particle sizes for that
class cannot be exceeded when tested. Refer to the ISO Clean Room Class chart (ISO 14644-1:1999) for values." I have looked at the maximum particle count limit for ISO 5 to ISO 8 and found that these numbers are per m3. But the unit in which VPC 300 is displaying is not clear to me.
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6384
  • Country: ca
  • Non-expert
Re: VPC300 particle counter
« Reply #4 on: March 24, 2023, 09:19:17 pm »
ug/m3 look like a unit of density but the measurements should be in terms of count per m3 or count per cm3, maybe ft3. I am not getting how it related to density.

If you have it set to cumulative, then its just counting the total number of particles. Does anything change if you set it to concentration mode?

But seems you are right, its counts per m3. Then if needed use some equations to convert it to ug https://www.researchgate.net/post/How_to_convert_mass_concentrations_ug_m3_to_particle_number_concentrations_particles_cm3
« Last Edit: March 24, 2023, 09:22:59 pm by thm_w »
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline joniengr081Topic starter

  • Regular Contributor
  • *
  • Posts: 159
  • Country: no
Re: VPC300 particle counter
« Reply #5 on: March 27, 2023, 09:11:56 am »
I understand the cumulative mode and differential mode but don't understand the concentration mode. How it differs then the other two modes ?

If it shows the numbers in the unit of per m3 than it's fine but it is not clearly written in the user manual. Where can I see the measurement unit documented ?
 