Thanks! The numbers for Z0 do seem a bit strange and I'll have to do some studying to make sure I understand what the results mean.
its explained in Agilent app note AN 8510-5B as transmission line's impedance after calibration plane. so i
assume:
1) its inside the cal kit, and it was assumed Zc (transmission line's impedance before calibration plane) = Zo = 50, when older VNAs dont support modeling it.
2) all offsets... Zo, delay, loss are physical, they can be measured with proper tools, the problem is there is no easy "proper tools", so now with advance computing, they are included in the curve fit modelling.
3) Steve's VNA support it, so thats why he included the parameter in the modelling..
sometime i wonder... sometime we can leave fringing capacitances parameters to all 0s and tune the offset delay (or with offset loss) alone to get "good enough" model, or set the delay to
0 some other value and Z0 = simply 50 ohm... and tune the C and/or L 0,1,2,3 for Open/Short to get similarly good enough model. so i guess whats important is the model parameters calculated by curve fit function. the actual physical offset value (Zo, delay, loss) is less important/impact/preferable (due to the difficulty of actually measuring them), as long as the model function used in the VNA (given the parameters calculated by the curve fit function) can give very close to the actual measured S11. with data-based VNA, curve fit is unecessary, only interpolation. ymmv.