Aliasing is more often an oscilloscope problem. A spectrum analyzer calls it the way it sees it.
The problem is "how it sees it!"
A transmitter on around 433MHz (in an ISM application, so it wasn't radiated to air) would usually work OK for several hours, then there would be sharp drop in output power from the transmitter.
Looking at a monitor point on the output, with the spectrum analyser (An early 2000s IFR), we saw that, when the fault happened, the single carrier at 433MHz would be replaced by multiple carriers spaced over a 40MHz range up to 477MHz.
Checking back to the carrier source, which was a PLL, gave a similar result.
The fault was annoyingly intermittent, we could get several hours of normal operation, then an episode.
The fault could adversely affect the process it was used for, so it was also important that the operator could tell when it happened, & call us, but leaving the IFR connected meant we couldn't use it in the workshop for other work, so we bought a little "El Cheapo" analog SA for the dedicated monitoring spot.
Luckily, one of the Techs was present the next time the fault occurred,.
The carrier smoothly slid up from 434MHz to around 477MHz----no spurious carrier or anything, just a PLL which had become unlocked!
The Tx fault was found to be that it required the data commanding the required frequency to be constantly available, & there was an intermittent fault in the header connecting this.
Data loss caused the PLL to go to its default frequency.
The IFR SA problem was that it didn't sweep continuously like the analog one, but, instead "swept" by taking samples over time.
This was perfectly good for most purposes, but a carrier shiftng fast in one direction was displayed as multiple carriers.
I guess you
could call the IFRs problem "aliasing."