Meter "Not calibrated"
But, the meter says directly "Not Calibrated".
Looks like the meter sees negative resistance and "resets calibration"
Tested this over again and calibration lost again.
Now I know, no problem, but want to tell this strange behaviour.
That's right, Karl-Heinz mentions this in detail in his manual on page 116 in section 5.5 SelfTest Function:
"Before the test steps begin, the zero resistance of the connected probes
is determined for all three combinations (T1:T3, T2:T3 and T1:T2). This zero resistances will be
subtracted for the future ESR and resistance measurements below 10Ω. Only zero resistance values
below 0.90Ω are accepted, because this correction values are not uses for measurement of resistor
values above 10Ω. If you use cables for measurement, you should only take cables with low resistance
values. If the later measured resistance results fall below the particular zero resistance for more than
0.2Ω, the tester will be resetted to ”uncalibrated”.
Madires in the m-firmware provided this option and made it possible to save and load 2 different calibration profiles.
For internal and external test contacts, which is very convenient to use. There is no need to carry out a new calibration every time.

For k-firmware there is no such possibility yet.
If someone (maybe me) was to turn the $20 tester circuit into a small PCB with a small OLED display, couldn't the Madires code be modified to run some pretty good tweezers?
Would you like to have such a tester device as in this video, I understand correctly?