The following fields are possibly (or certainly) truncated
IAP Start mode:
Cal Mode Start:
Cal Mode Key:
Please follow the detailed descriptions for these three items in the chapters: Calibration Mode, Zero Offset Calibration, Full Scale Gain Calibration and Firmware Update.
LowZ/R1
".. are not available"
DCV and ACV/R2,R3
That's 500.00V and 600.0V in both cases. These are the necessary test signals for full scale calibration only.
PeakHold/R0
".. at AC5V Range". Sorry, I can't describe, what's exactly necessary here, as I did never calibrate this mode. Maybe it's not necessary, as zero offset does not show up in the according chapter for this mode.
Capacitor/OFFSET
Here again, I can't tell, what's to be done here, as I never tried that calibration. This mode also does not show up in the zero offset chapter.
I speculate, that you need to calibrate zero only in R1 = 100nF range, with open jacks, or if you want to include your standard test cables, with latter left open.
The 50MOhm range is also calibrated at 50000 counts, i.e. 50MOhm, not @ 40MOhm.
Just another general remark. The calibration of this DMM (as well as other 5000 / 6000 count DMMs) require special calibrators which provides these specific F.S. signals for resistance and capacitance, instead of the usual fixed values like 10000 / 19000 Ohm.
That is done by a method called synthesized impedance, i.e. which outputs fully variable calibration stimuli, as in the FLUKE calibrator 5500A.
Frank