| joeqsmith:
I remember the GMW being marketed as an industrial meter which to me suggests 10V/m, criterion A. Looking through the standard, it appears that all three level (1,3 & 10) require performance criterion A. But looking closer, at the definition of A, "No degradation of performance or loss of function is allowed below a performance level specified by the manufacturer, when the equipment is used as intended." which backs up your comment about Brymen. When I swept those four meters, the GMW was by far the most effected, followed by the UNI-T UT181A. *** Should add, that was with the meters terminated to 50ohms, no test leads and set to their lowest DCV scale. |