121GW Voltage Readings WAY WAY Off.   Not a 121GW general issue. Moderator, if you think it is move this message please.

I was trying to trouble-shoot with my 121GW.  It hadn't used in several months.  I was checking out a few on-board power supplies of an instrument I am trying to check.  The 6 volt supply was reading 8.444.  I trouble-shot that for an hour then I used my Agilent DVM and it read 6.001 volts!  Son of a bxxch!!!

I've never used the 121GW for voltage measurements, just ohms and didn't like it very much for that and really just set it aside after that.  Now this.  I am using the exact same set of meter leads by the way, I just moved them between DVMs.

I think it may be time to give this 121GW a toss.  It never has been right since I bought it and now, I suspect it is over a year old now.

Dave
NoisyBoy:
With reading this far off, the first thing I would check is to put in a fresh set of batteries.
ggchab:
The position of the rotary switch is not always perfectly correct. I had the problem with the resistors'range a few times and it gave incorrect readings.
rsjsouza:
Does teh voltage have AC components as well? If so, check:
https://www.eevblog.com/forum/testgear/eevblog-121gw-multimeter-issues/msg3015706/?topicseen#msg3015706

Otherwise, let me know if you plan to toss it... I may be interested. :-+
dcac:
Also check if the error persists if you change range.