Products => Test Equipment => Topic started by: quarks on April 14, 2014, 04:32:50 pm
-
Hello,
I read about that ERROR in the PMEL Forum
see
http://www.pmelforum.com/index.php?topic=1070.msg20940#msg20940 (http://www.pmelforum.com/index.php?topic=1070.msg20940#msg20940)
I wonder what msrichmond found out when he stated: "Never mind, something simple...timing oscillator is not timing."
I registered at PMEL and send him a PM and also asked if anyone there knows the solution for the described problem, but unfortunately got no response.
Does by chance anyone here know what the solution for the problem was?
thanks
quarks
-
unfortunately so far nobody here and in pmel forum replied.
My investigations on this error showed some strange behavior.
When in DCV mode with long integration time the reading is freezing after a short time (sometimes almost instantly and sometimes between 1s to 15s). With very fast sampling time (<NPLC 0.01) it is running without problem. Also in ACV it is not showing the error.
Does anyone have a clue/hint/idea?
thanks
quarks
-
Does anyone have a clue/hint/idea?
Howdy! With the precision of a 3458A..meter, Agilent and $2740.46 comes to mind. I just sent one to the Loveland, Colorado repair lab and it cost a nice $2740.46 .. no questions asked. They went through the entire meter replacing any and everything that even looked worn or broken.. including the keypad, outer bezel, rear panel and display , this includes a max/min / 2 cal which is the factory spec when the meter leaves the factory. I opted for the Standards Lab cal on top of that because the meter really deserves a really good calibration using gold standards. The service experience was awesome as well.. totally white glove and very professional.
Just a suggestion if the resources are available..
JLM
-
this is really just a guess, as you would expect based on the wideness of the problem mentioned. what came in mind is something i saw on an other hp test gear. if the relay switching the voltage to the a/d does not work properly (i.e. contact closure bad/instable) the voltage the converter sees may have fluctuations over time, these may be covered in short conversion cycles, but in long cycles the multislope converter may just not be able to converge and stop. again, just one potential answer to the question.
but please, if you find out, let us know
-
Does anyone have a clue/hint/idea?
Howdy! With the precision of a 3458A..meter, Agilent and $2740.46 comes to mind. I just sent one to the Loveland, Colorado repair lab and it cost a nice $2740.46 .. no questions asked. They went through the entire meter replacing any and everything that even looked worn or broken.. including the keypad, outer bezel, rear panel and display , this includes a max/min / 2 cal which is the factory spec when the meter leaves the factory. I opted for the Standards Lab cal on top of that because the meter really deserves a really good calibration using gold standards. The service experience was awesome as well.. totally white glove and very professional.
Just a suggestion if the resources are available..
JLM
Thanks for the reply. I know/knew about this option and if I could get your described "Loveland Service" it really would be great. But unfortunately I am in Germany and I do not know if the German Service is able to provide the same level as Loveland (my guess is probably not). Also I want try to find the root cause. But if I fail, I will remember your suggestion.
this is really just a guess, as you would expect based on the wideness of the problem mentioned. what came in mind is something i saw on an other hp test gear. if the relay switching the voltage to the a/d does not work properly (i.e. contact closure bad/instable) the voltage the converter sees may have fluctuations over time, these may be covered in short conversion cycles, but in long cycles the multislope converter may just not be able to converge and stop. again, just one potential answer to the question.
but please, if you find out, let us know
thanks for your reply. Right now I still think/hope it is something really simple as the PMEL poster seemed to have found out. So I will further investigate.
-
So by reading the blog you linked it seem that what the user would say that the internal instrument frequency is not right. Now If that user change the A/D board and nothingh change we are sure that the problem is somewhere else. A few time ago I saw a schematics of the device.
The unit use an inetrnal 10Mhz reference so If it is bad you could have that error.
Best regards, Alberto.
-
Quarks,,
oh sorry, I wasn't aware you were in Germany.. sometimes I forget there are other places these things could exist. .. :)
I know this is a long shot without the service / functional operation in front of me. Do you know for sure if your calibration constant NVRAM is in good shape? Is if possible the battery has died in that chip causing this mode of failure? Do you get any other errors when you first power the machine?
JLM
-
Problem is only on A3 board. Nothing else is involved. Therefore unfortunately the PMEL threat was leading me in a wrong direction. A3 board swap fixed the problem.
-
so have you found out what failure it was then on the a3 board. have same failure.
-
never mind, fixed it, it was on the a/d converter board. seems to be a systematic problem as so many show this error.
-
What did you fix?
-
I have device with the same error.
Further investigation give me result that trigger signal (LEVEL) from A2 AC board is entered to pin 11 of U210 on A3 board, but signal SHTRG (XT/H) which must drive Sample and Hold on A2 AC board is not going out. Suspected is U210 or missing signal. I'm looking for detailed description of operation of A3 board or better U210 CMOS array. Also looking for defective A3 board with good U210 or only U210 (1820-5770) MB651314 from Fujitsu.
I would be grateful for any help.
-
I found the problem with help from frank_stellmach from Yahoo Agilent group (Thank you Frank) - it was in faulty U405 EL2018. Hope in the future someone can use this info.
-
I found the problem with help from frank_stellmach from Yahoo Agilent group (Thank you Frank) - it was in faulty U405 EL2018. Hope in the future someone can use this info.
The would be Dr. Frank on this forum
https://www.eevblog.com/forum/profile/?u=53397 (https://www.eevblog.com/forum/profile/?u=53397)
/Bingo
-
Yes, I'm the same person.
Anyhow.
This 3458A would be rescued, only if I'm really correct in my assumption, and if an appropriate solution can be found and implemented.
These obsolete EL2018 (~5 EA in the 3458A) seem to fail after many years.
Several ADC errors also have their root cause to their breakdown.
I've seen a picture of an agilent original (?) replacement of these comparators on the ADC board, by an SMD two fold fast comparator, (piggy back) but I cannot remember the correct site, nor supplier, nor type designator, I'm so sorry.
Maybe anybody else can help.
Otherwise, give the Chinese sellers a chance.
Frank
-
Dr Frank,
I bought a board from Agilent last year. It has the mod to which you most likely refer. I have attached a couple of full size images. It appears to use the EL2252CN EL2252CM comparator. I can try to get better pics if anyone is interested.
-
ManateeMafia,
your pictures are very good, but the outlook for defective 3458As are not..
the replacement EL2252 is also obsolete, as it has been manufactured by the same supplier.
Frank
-
ManateeMafia,
your pictures are very good, but the outlook for defective 3458As are not..
the replacement EL2252 is also obsolete, as it has been manufactured by the same supplier.
Frank
Rochester still has some.
With best regards
Andreas
-
Ebay is filled with EL2252s
http://www.ebay.com/sch/i.html?_odkw=el+2018&_osacat=0&_from=R40&_trksid=p2045573.m570.l1313.TR0.TRC0.H0.XEL2252.TRS0&_nkw=EL2252&_sacat=0 (http://www.ebay.com/sch/i.html?_odkw=el+2018&_osacat=0&_from=R40&_trksid=p2045573.m570.l1313.TR0.TRC0.H0.XEL2252.TRS0&_nkw=EL2252&_sacat=0)
Take these:
http://www.ebay.com/itm/QTY-5-EL2252CN-ELANTEC-14-PIN-DIP-VOLTAGE-COMPARATOR-FREE-SHIPPING-/121203386865?pt=LH_DefaultDomain_0&hash=item1c3848e9f1 (http://www.ebay.com/itm/QTY-5-EL2252CN-ELANTEC-14-PIN-DIP-VOLTAGE-COMPARATOR-FREE-SHIPPING-/121203386865?pt=LH_DefaultDomain_0&hash=item1c3848e9f1)
Take cheap DIP8 IC (NE555), a dremel and drill it's guts (die) out. Then shorten/cut the leads of the EL2252, stick it with some super glue on the back of the gutted NE555 and use some thin copper wire to rewire one of the EL2018 inside EL2252 to the top of the NE555 pins.
-
Few days ago receive EL2018 from HKUTSource http://www.ebay.com/itm/EL2018CN-Encapsulation-DIP-8-Fast-/380658308131?pt=LH_DefaultDomain_0&hash=item58a100d423 (http://www.ebay.com/itm/EL2018CN-Encapsulation-DIP-8-Fast-/380658308131?pt=LH_DefaultDomain_0&hash=item58a100d423)
Replace U405 on device and tadaaa - now 3458A have ACSyncro working fine. As already write issue was confirmed by replacing this chip with other comparator, but now working how HP designed.
I want to tell one more time Thank you to Dr. Frank for guiding in right direction.
-
Hi to all,
I found this forum ,having errors 209 & 210. No measurements in VDC,ADC, Ohms & Freqs. ACAL impossible. I tried to replace 2x EL2018CN, don't work better.
I see that U405 might be replaced, but many talk of a EL2018CN but on my 3458A it's an MC34072P, which are not the same.
Some are aware of those differences?
Thanks for any help,
Damien.
-
U405 on the A2 board is the MC34072P.
U405 on the A3 board is the EL2018.
Change NPLC to .001 and see if you can at least get some readings.
-
Thank you very much for your help Manatee Mafia.
Effectively it was a confusion from me . I searched U405 for long, but i have the EL2252CM modification in place of U142 & U405. Replaced by two EL2018CN and it now works fine, got readings @ NPLC 0.001 .
Now i have the "multislope or balanced" convergence error, seems U301 A3's defective, got to replace the R-2521 receiver...
-
broketest
Now i have the "multislope or balanced" convergence error, seems U301 A3's defective, got to replace the R-2521 receiver...
Bad news, this error in 99.8% means charge slope hybrid is gonesky, and you are in for new A3 replacement (1.3K$ from Keysight). If optical intereface bad it would just throw ISOLATOR FAILURE.
-
Hi to all,
I come back after this 3458A that still give me hard times. After replacing the A/D board by a fresh one from Keysight, the device now reads in DC, pass the Autocal but totally randomly fails the autocal @ ACV Ratio 1, or even while using indicates Overload on AC channel, and no reading anymore in Ohms too.
A tech @ Keysight told me that my 3458a is an older version and needs to be updated with the newest ROM, which is way too expensive to add to the costs already put in.
I suspect a temperature drift or error somewhere... Anyone have an idea? Thanks in advance for any tip :)
-
What is it reading in AC 10V mode? Is it something unreaslistically high like 40V?
That would be an indication the Elantec IC is bad. I would have to look up its U number but it is common.
-
Hi to all,
..
A tech @ Keysight told me that my 3458a is an older version and needs to be updated with the newest ROM, which is way too expensive to add to the costs already put in.
I suspect a temperature drift or error somewhere... Anyone have an idea? Thanks in advance for any tip :)
Please identify your current ROM version, and the number of EPROMs inside your unit, very probably, it's a 6 IC version.
You can program the latest version, 8 or 9 ( no difference) by yourself, at no cost, if you erase the existing EPROMs and reprogram them.
There exist different versions of the firmware in the net, and somebody already had partitioned the newer 1 - chip -versions into 6 files .
All firmware versions will work with your board, and a re-calibration is also not necessary.
This will not cure your ACV problem, but will fix a number of bugs.
You might also equip your instrument with opt. 01, 128k additional memory, by adding 4 x32k RAMs.
The replacement of the CALRAM is also recommended, if its datecode it elder than 10 years.
Frank.
-
The replacement of the CALRAM is also recommended, if its datecode it elder than 10 years.
Frank.
Before mucking around with the memories, I HIGHLY recommend making a backup copy of the CALRAM data. I wrote a program (based on Poul-Henning Kamp's work) to do this over GPIB. The backup program is now included in ke5fx's GPIBKIT package. Note that you cannot write the calram over GPIB... you will need a device programmer to do that. Once the calram chip is pulled, I'd also back it up using the device programmer... belts and suspenders are a good thing.
-
Hi all,
I am aware this is an old thread, but I ran into the "Error 209: Hardware Failure: Timeout - Unable to read A/D" when replacing an older U210 with a newer one.
Interestingly, the ACAL works on everything but AC. Also, there are no measurements, even when setting NPLC to 0.001.
This is what the U210 chips look like:
https://xdevs.com/doc/HP_Agilent_Keysight/3458A/a3pcbs/pcba_a3_d.jpg (https://xdevs.com/doc/HP_Agilent_Keysight/3458A/a3pcbs/pcba_a3_d.jpg) [original]
https://xdevs.com/doc/HP_Agilent_Keysight/3458A/a3pcbs/pcba_a3_n.jpg (https://xdevs.com/doc/HP_Agilent_Keysight/3458A/a3pcbs/pcba_a3_n.jpg) [replacement]
The reason for attempting a replacement is that the room when to 35 C one day in the summer and failed unexpectedly within the 3458A operating range. Now the drift is all over the place. I have hope that it is something else on the board besides the U180.
On replacement of U210, it appears that I need the daughter board (03458-66530) that is connected to U142 and U405. Luckily, someone has already reproduce the board for us on PCBWay:
https://www.pcbway.com/project/shareproject/EL2018_EL2252_board_for_HP_3458A_b625a37e.html (https://www.pcbway.com/project/shareproject/EL2018_EL2252_board_for_HP_3458A_b625a37e.html)
Has anyone 'uprgraded' their U210 before? What other board changes were made besides the daughter board? I do not see anything else significantly different.
Regards.
-
U210 in the original is a gate array chip with digital logic. So it is not really a cadidate to have issue with a slightly higher temperature. I would consider a replacement there an odd attempt to fix a problem with high temperature. The replacements are some CPLDs/FPGAs with level shifters. As U210 replacements are rare this is more a change at production. Ideally there should be no extra changes needed, but not all version may work mechanically with all PCB versions.
The adapter board is not related and a replacement for the fast comperator with a dual version, that is slightly better available.
-
I replaced the two E2018 chips with the E2252 daughter board. Previously, the U210 chip was swapped.
The symptoms are the same:
There are sometimes readings on startup.
202, "Hardware Failure -- Slave Test: Convergence"
Test fails.
ACAL DC pass
ACAL OHMS pass
ACAL AC (instantly fails)
209, "Hardware Failure -- Timeout: Unable to Read A/D: 243"
101, "Calibration Error"
I tested the supply voltages for the A2 (AC) board:
+2V ... 2.5V +/- 1.2mV
+5Va ... 4.6V [questionable]
+18V ... 18.4V +/- 0.4mV
+14Va ... 14.0V
-18V ... -18.4V +/- 0.6mV
-15Vd ... 15.3V
-17Va ... 17.6V
Seems OK.
What I have noticed is that the board occasionally works when it is cold. I am wondering if a part or a solder joint could be bad somewhere.