Hi,
I've just received a HDM3065. Quite pleased with it so far, except... I've noticed a bit of a problem.
Since I temporarily have an Agilent 34461A on my bench, I decided to measure the same voltage source (~29V) at the same time and see how the two tracked. Pretty good until ~36 minutes into the test the HDM suddenly showed a measurement of ~53V. And then every 36 minutes thereafter, the same thing. Roughly 60V, give or take, rogue reading. Makes a bit of a mockery of the stats! The spikes are very definitely not real, as shown by the Agilent and also tested with a 'scope.
Attached are two screenshots showing the issue. Note the spike to the right of the Hantek graph.
Same thing on current measurement, spikes every ~36 minutes except -ve going this time showing as '-ve overload', and 2W resistance - once again higher readings than is the truth.
This smells of software to me - but I can't seem to find any updated firmware on Hantek's website. I see there is a thread on SW for this meter but haven't had time to digest it and my skim of it filled me with dread...
The software version is 2.0.0.6(2394081348), firmware is 04.01.07.04.23-02.05-21.
Has anyone else noticed this issue? Or do I have a unit with a pretty whacky fault?
I've emailed support@hantek.com but received no reply yet. This, I understand, is to be expected.

Such a shame because the hardware seems pretty nice.
Cheers,
Jez.