7265 Lock In Amplifier

<< < (3/6) > >>

_Wim_:
Just gave it a quick try. It discovers my 7260 (connected via an Agilent 82357b GPIB adaptor). But when I try to open the device, I get time-out errors.

_Wim_:
Maybe just a quick question, does it discover all instruments, or only specifically look for 726x instruments? I am also not sure how to interprete the discoveries. My 7260 is at GPIB 12, but all discoveries are listed as "0"...

Edit:
All devices are discovered, and the 7260 is the one indicated with all the zero's, as the 2 other names are 2 keithley devices on the bus. But as explained, when trying to connect, I get a time-out for some reason. 

_Wim_:
I can connect via the same GPIB adaptor using the acquire software. I did have to reboot the 7260, as it was no longer detected by both your software or the Acquire software.

As you wrote you are using the same GPIB adaptor and could not connect via acquire, I suspect our adaptors are configured differently.

_Wim_:
Just tried by enabling "488" modus, but this did not do the trick. I can still connect via acquire, but still get a time-out with your software.

_Wim_:
Some additional info:

When I send ID via connection expert, I get '7260' back. This is also visible in the "comm test" menu of the 7260.

When I open your app, during the browsing of the devices I see the addition in the "comm test output" of [LF][EOI][LF][EOI]. When trying to opening the device, nothing additional is seen in the comm test window.