In another thread, I stated that my Fluke 17B+ was way of calibration spec., specially when compared to the results of a Fluke 117. When I contacted Fluke's support, I said that I was trying to read the output of a voltage reference, be it a MAX6126AASA41+ or a MAX6126AASA50+, or even a ISL21009DFB841Z (4.096 Vref from Intersil, which is what is all I have left on hands right now: the MAX6126 are gone). They told me that the difference in reading was expected, because the Fluke 17B+ would only properly read resistive voltages, and a voltage reference was a capacitive voltage. I had no was to refute that argument because I didn't know if the guy was right or wrong.
Anyhow, I ran more tests on a voltage divider using two 10k Ohm 2% resistors, measured to be within expected accuracy. According to the Fluke tech support guy, this should be enough.
I also ran tests using a variable PSU based on a LM317 and the same voltage divider.
According to
http://www.fluke.com/fluke/brpt/multimetros-digitais/general-purpose-multimeters/fluke-17b+.htm?pid=78682, the specs for DC for the 17B+ are:
| Function | Range | Resolution | Precision |
| DC Voltage | 4,000 V 40,00 V 400,0 V 1000 V | 0,001 V 0,01 V 0,1 V 1 V | 0,5% + 3 |
In my tests, I also compared it to a cheap Mastech MS8260E (bought from DealExtreme several years ago). The funny thing is that in all the tests, the Mastech's results were always on pair with the Fluke 117.
Here are the results I got:
Output from a ISL21009DFB841Z:
| Method | Fluke 17B+ | Fluke 117 | Mastech 8260E |
| Direct reading | 4.050 | 4.097 | 4.09 |
| 10k/10k voltage divider | 2.020 | 2.044 | 2.03 |
In the second part of the test, I used a variable PSU based on a LM317, and used the Mastech 8260E to read its direct output. The LM317's output was also connected to the same voltage divider as above, and half of the divider was read by both the Fluke 17B+ and the 117. My expectation was to have the Flukes always give 50% of the reading of the Mastech. This is what I got:
| Mastech MS8260E | Fluke 117 | Fluke 17B+ |
| 1.764 | 0.877 | 0.866 |
| 4.00 | 1.999 | 1.975 |
| 5.00 | 2.498 | 2.469 |
| 6.00 | 2.999 | 2.963 |
| 7.00 | 3.495 | 3.454 |
| 8.00 | 3.992 | 3.945 |
| 9.00 | 4.493 | 4.44 |
| 10.01 | 4.995 | 4.93 |
| 10.63 | 5.301 | 5.24 |
Conclusion: the Fluke 117 and the MS8260E are giving compatible results, but I am not sure if I should pursue an RMA for the 17B+, and this is what I need your help with: is it really out of spec and, if so, is the table above enought argument to convince fluke that it is out of spec?
thanks in advance.