| vaualbus:
so you suggest to use normal eletroliti caps in the 33120A? All this caps are bypass capicitor on the -18V and 18v rail? If it is possible it even better for so I haven't to buy the caps. |
| vaualbus:
I've some update for you guys: Fist I replaced all the caps and diode and now the display not work (The fronpanel fall from my desk so the vacum display could be broken but I don't see any crack), but the instrument still not produce any output. The frontpanel a part from the display it work, I can move the encoder an hear the click, use buttons and ecc. Probably I find some problem. The negative +15V read -20 meanwhile the positive +15V read only 9V. I don't kno why I read this voltages? PS Somebody have some frontpanel or vacum display to sell to me? |
| vaualbus:
I was wrong aboat the volatages I mesearud them with the wrong GND. Both 15 an -15 are right. |
| vaualbus:
Also how I can meseuare the output of the dac? It seem a stupid question but it isn't the dac have a differential current output so I can't meseare that with a scope. Any idea? |
| TiN:
Check getter on VFD. If there is white residue around getter, it means oxygen got in and vacuum is gone now. If its metallic shiny, your display likely be okay. |