Products > Test Equipment
A High-Performance Open Source Oscilloscope: development log & future ideas
tom66:
Does raise the question of how to use the 12-bit 640MSa/s mode. I had considered limiting that to 500MSa/s as that fits in an even multiple of 16-bit samples with 4 bits unused. But that makes the 4ch Nyquist 62.5MHz and if you have a 12 bit ADC with ~10.5 ENOB after AFE noise, then you need filter to be rolling off to -63dB with say a -3dB bandwidth of 40-50MHz. Even enabling the full 640MSa/s is still only 80MHz Nyquist so practical upper B/W limit is still ca 50MHz.
I don't think that is practical so 12-bit mode will always have some risk of aliasing if used on 4ch mode. Switching filters (other than a simple 20MHz Varicap filter) seems impractical and in any case a single pole filter driven by a varicap is unlikely to roll off quickly enough to be useful for 12-bit mode.
So what do you do?
Marco:
--- Quote from: rhb on December 11, 2020, 02:05:16 am ---My primary concern now is the AFE input filter. It should be a high order Bessel-Thomson filter to provide accurate waveform shape. I've got every reference I can find, but unfortunately, the maximally flat phase gets skimpy treatment and I've still not figured out how to analyze and design one from first principles.
--- End quote ---
I know nothing about it but what google told me and a history of stuff from college I forgot after the exam, but I found the Incomplete Gaussian Filter from this paper rather elegant. Looks easy to implement and doesn't try to reflect a ton of energy into your buffer.
rhb:
--- Quote from: nctnico on December 13, 2020, 09:08:52 pm ---
Just think about it and look at it from a practical point of view. Frequency continues to roll off, signals consist of harmonics and at 200MHz you are already over the limit of what can be measured with a standard hi-impedance probe. The probe itself will already cause a significant high frequency attenuation.
There is a ton of information available on this forum about anti-aliasing filters and DSOs. But since this thread is about an open source design you are free to fit whatever filter you like. I will go for what is the standard approach (which is to have a bandwidth of fs/2.5) for now.
In a nutshell:
From an error perspective: 1% is more than 2 bits (2 bits = 12dB). So if the attenuation is 3dB at 0.4fs, 48 - 12 = 36dB at Nyquist (0.5 fs) and 48dB at 0.6 fs then the amplitude error is less than 1% due to aliasing. Another issue to factor in is that in order to show the shape of a waveform you will at the very least want to see the first 2 (base and 1st) and preferably at least 3 of the harmonic frequencies. For an aliasing error to occur a harmonic frequency would need to be between 0.5 fs and 0.6 fs (and be closer to .5 fs to have the biggest impact). Remember that an oscilloscope isn't a precision instrument nor a data acquisition device and at the -3dB point the amplitude error is already near 30% !
In the end it is all about compromises; getting the highest bandwidth with the least horrible step response. And there is always the option to include two filters; one with the best step response and one with the highest bandwidth.
--- End quote ---
Here is the log magnitude spectrum of Bessel filters from order 1 to 10 from .
Handbook of Filter Synthesis
Anatol I. Zverev
Wiley 1967 & 2005
As can be seen from the attached figure, with Fc = 200 MHz, at 250 MHz there is less than -6 dB of attenuation even for a 10th order Bessel filter. At 1 GHz, a 5th order Bessel is down about -50 dB.
With that I shall leave you to figure out what you got wrong.
Have Fun!
Reg
nctnico:
@rhb: There must be something wrong with the tables you are using. The filter generator & simulator tool I'm using shows a decent attenuation at Nyquist. Unfortunately I'm on the road so I don't have access to it so that has to do for now.
--- Quote from: tom66 on December 14, 2020, 05:24:52 pm ---Does raise the question of how to use the 12-bit 640MSa/s mode. I had considered limiting that to 500MSa/s as that fits in an even multiple of 16-bit samples with 4 bits unused. But that makes the 4ch Nyquist 62.5MHz and if you have a 12 bit ADC with ~10.5 ENOB after AFE noise, then you need filter to be rolling off to -63dB with say a -3dB bandwidth of 40-50MHz. Even enabling the full 640MSa/s is still only 80MHz Nyquist so practical upper B/W limit is still ca 50MHz.
I don't think that is practical so 12-bit mode will always have some risk of aliasing if used on 4ch mode. Switching filters (other than a simple 20MHz Varicap filter) seems impractical and in any case a single pole filter driven by a varicap is unlikely to roll off quickly enough to be useful for 12-bit mode.
So what do you do?
--- End quote ---
I have been thinking about this a bit. I think a good approach would be to have several filter banks and use a mux to switch between them to select a different roll-off for a different bit-width and maximum samplerate. A higher order filter using passive components is not very difficult and not very expensive to implement.
tom66:
--- Quote from: nctnico on December 14, 2020, 09:47:53 pm ---I have been thinking about this a bit. I think a good approach would be to have several filter banks and use a mux to switch between them to select a different roll-off for a different bit-width and maximum samplerate. A higher order filter using passive components is not very difficult and not very expensive to implement.
--- End quote ---
It is quite expensive to do for 100MHz+ and across 4 channels, though. Remember, these things also need to be tested during production with a sweep generator, and possibly need manual adjustment. There are a few manual adjustment varicap points on even cheap oscilloscopes, which appears to be for matching input capacitance. I'd really like to avoid doing that with the filters, and I don't think there's the grunt to do real time DSP on the ADC samples and correct AFE response there (unless it was very limited in response, and ~all the DSP blocks were used.)
I have tested my Rigol DS1000Z, and it aliases in 4 channel mode. I think it's just something that's rather difficult to avoid, and the operator just needs to be careful not to exceed the parameters of their instrument. If the measurement is really crucial, then an external analog filter could be used.
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version