Docsaad:
No i did not backup the eeprom . I had no idea it would do this. Otherwise I would have never done that |O. Thank you for your tip on msging @indman. Lets hope he has a solution.
indman:
I have specially selected with big font
ATTENTION! To do only if there is a backup copy of EEPROM!
If you haven't made it, then it will be very difficult to restore factory settings. I can offer only an firmware dump from other UT210E. Try to update an firmware this file. I have no other decision.
agaelema:
What range lost the calibration? I think a better idea is to use the same process and re-calibrate the meter.
Docsaad:
Voltage in dc amps in dc. Basically all the useful calibrations went off
agaelema:
Normally the calibration affects just the selected scale.
In this case will be necessary a variable load and another meter as a reference. If I'm not wrong, a short press of hold button decreases the value and a long press increase the value.