@auato
It is possible there have been some small piece of what ever inside attenuator relay side.
But in more worst case there can be some other problem inside attenuator cave. I do not list all what are somehow possible.
I think your some one previous images was not from your attenuator but similar (because it is partially disassembled and bit weird way, perhaps just for image...who knows). I have used this image (part of it) here.

There need always carefully look these positions are clean (marked red arrows). These metal surfaces need be free from some small garbage and also these surfaces of nylon pins, so that it can move perfectly full amount. (I have seen many times example small piece of old brittle O-ring rubber on these surfaces, sometimes these are like attached there and need carefully remove if they exist and if still do not yet want do whole renew work (there is two methods for take nylon needles out from contact strips in cave - very risky "lottery" way or very safe but lot of more "watchmaker" work. (example I use always dust free air room and and magnifier (microscope in my case dye to my after "best before" aged eyes )
Btw, in some pictures in this thread can see some attenuators where can see bit weird contact strips and also really weird contact strips positions... without further explanations and other images can not tell what are these and why... there is one image where strips are positions I have never seen when nylon needles have removed... really weird. (but because it is not your attenuator ... no need think more these)
If you use compressed air to inside attenuator cave. Be very careful because these thin contact strips. They are fragile. Some times also old glue can brittle (they are glued in perfect positions). Also air need be real clean and oil free, and still it can do more bad than good because some particles attached to the surfaces inside may detach and then remain in contact. The low contact pressure contact is to some extent "self-cleaning", however, because there is a small slip when the contact is made... gold alloy against gold alloy.
If the cavity of the attenuator and the end flanges have never been opened, it is quite unlikely that there is debris there. The holes of the nylon needles have quite a small clearance, so normally no debris gets in there.
But what have I ever found. Rarely. Some kind of "oil" or "grease". It is put in when the witches are made. There, at the point of contact between the nylon needles and the contact strips. Usually it is a very small dose. And it must never come into contact with the surface. Sometimes the attenuators run very hot (production test use, for example, is very tough, the user's HP didn't even give them a warranty for that use).
In such a situation, the fat may heat up and drain, especially if there has been a bit too much of it there. Grease between the contact surface does not prevent contact. But then the typical problem is that during the SelfCal / Self Check phase the contact is created a little too slowly and it fails or sometimes even gives the wrong value (capasitive contact just before galvanic)