If you are using them AM503 amplifier with that probe
He must use the AM503, when using the A6303.
Scope settings are once done, everything else should be done on the AM503.(Except timebase/triggerlevel, of course)
Scope is set to 10mV/Div,
DC coupling, 50 ohm termination, no bandwithlimiter on.
Bandwithlimit, dc or ac coupling, this will be done at the AM503 module.
Or like the manual "say":
NOTE. To maintain accurate readings while using the AM 503B and AM 5030,
the vertical gain of the oscilloscope channel must always remain at 10 mV/Div
and the coupling must remain at DC
Further the scope should have five times the bandwith, the probe got.
In case of the 6303 this shouldn´t be a problem...