Products > Test Equipment
Adjusting of coupled filters with a VNA
(1/1)
DeepLink:
Hi
Need to adjust some coupled filters, (3 to 5 poles - below 500MHz)
Having read the app note 1287-8 from HP, it seems that "Time Domain Analysis" is the way to go
I do have
HP 8714ES w. opt. 100
Siglent SVA1032X w. DTF
None of these VNA's offer real Time Domain Analysis, but are the DTF options any good for adjusting filters?
I can't afford a E5071 or other fancy equipment, so I'm looking for a way to do it with what I have now, or the next step up
Maybe a HP 8753* with opt. 010?
How are you adjusting coupled filters?
S57UUU:
I guess these are bandpasses?
You can tune them in frequency domain, with a normal network analyzer (even a scalar one).
Measure S11 (do not forget a load at port 2).
Tune all of the resonators way off (or short them). Do a response calibration (normalization).
Tune the first resonator, to get a minimum in S11 at center frequency. At this stage set the scale to 1dB/div or less, since the minimum will be very shallow. You can deepen it by introducing some loss (finger...) into resonator #2.
Next, tune resonator #2. Depending on your coupling between the resonators (filter width), the minimum will split into two, or just become deeper. If split, tune #2 so, that the two dips are symmetrical around center frequency.
Tune #3 to get three symmetrical dips, etc.
Marko Cebokli
Navigation
[0] Message Index
Go to full version