Hi there. I have a R3671 and it is reading about 1.2 dBm low across all ranges. Do you have the service password for the unit and how much do you think is needed to correct for an error like this? When I try to do a Cal, I get the error IF Input Level Out of Range [-2544].

Warm Regards,
-- Mark
suj:
Try the password from this nctnico post:
https://www.eevblog.com/forum/testgear/advantest-r3477-9khz-to-13ghz-spectrum-analyzer/msg1004897/#msg1004897
It's working for my R3681
Mgelderland:
Hoi nctnico
Is it possible to get a copy of your flash drive
I have a unit without a flash drive
Perhaps it is possible to get it working again this way

with kind regards

Marlon
nctnico:

--- Quote from: Mgelderland on October 13, 2024, 04:50:30 pm ---Hoi nctnico
Is it possible to get a copy of your flash drive
I have a unit without a flash drive
Perhaps it is possible to get it working again this way

--- End quote ---
I have an image made using Linux with the dd command so this is a raw disk image. IIRC I have tested whether it works in my machine. You'll need to use a >=512 compact flash card as the image is 512MB. If you PM me your email address, I can send the image through Wetransfer. I have no idea though whether licenses and so on are tied to the serial number of the machine.
Mgelderland:
thank you very much
I transferred the files to the flash drive and it works again.
what options were there on your device?
for me (23 50) are these software options or hardware?