I will update the topic.
I received such a network analyzer and planned to use it instead of my R3754A.
I don't have much experience with such devices, but it seems to me that the 3753 does not work properly.
I had to repair the display and replace the capacitors (almost all 33uF25V degraded), but this did not change the operation of the analyzer.
I turned on the analyzer, connected it with wires:
Output 2 and input R
Output 2 and input A
r3753_connection.png
I see on the screen from time to time flashing the inscription "Rch Overload"
adv1.jpg
I figured out that the problem is with the transmitter.
You can notice spikes on the waveform, which is why I see the overload message.
adv2.jpg
adv5.jpg
If I do not set the full range (5Hz-500MHz), but with 1MHz, then the problem goes away.
adv3.jpg
The problem does not appear in the low frequency range of 5 Hz - 999.999 kHz
adv4.jpg
I found out that this is a problem of switching from 999.9999KHz to 1MHz
adv6.jpg
adv7.jpg
I don't have any ideas yet.
Does anyone have such a device?