Agilent 34401A 6.5 Digit MM Problem
« on: June 06, 2023, 02:46:22 am »
Hello everyone,

I'm new to the group, so first post.  I bought a really nice Agilent 34401A on ebay, and i have a little issue i need help with.  The meter works great, other than the decimal digits blink really fast.  more like jitters.  And when you power it up, it displays a quick message "addr 22" or something like that.  can anyone help with suggestions?

Thank you kindly
 

Offline Shonky

  • Frequent Contributor
  • **
  • Posts: 290
  • Country: au
Re: Agilent 34401A 6.5 Digit MM Problem
« Reply #1 on: June 06, 2023, 03:50:21 am »
Addr 22 is just the programmed HP-IB (GPIB) address. Totally normal.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5884
  • Country: ca
Re: Agilent 34401A 6.5 Digit MM Problem
« Reply #2 on: June 06, 2023, 06:32:50 am »
yes  it shows it gpib address

please  read the user manual(s) who can be found on the web

you'll learn about a few functions and or settings
 

Offline AVGresponding

  • Super Contributor
  • ***
  • Posts: 4660
  • Country: england
  • Exploring Rabbit Holes Since The 1970s
Re: Agilent 34401A 6.5 Digit MM Problem
« Reply #3 on: June 06, 2023, 08:58:20 am »
There are a lot of features and settings to learn about on these, and to get the most out of it, you really need that GPIB to connect it to a PC.

The flickering digits are due to the high update rate it'll be set on. You can lower it in the menu system, which you can access on the front panel by pressing "shift" then the left arrow, and use the arrow keys to navigate around.
nuqDaq yuch Dapol?
Addiction count: Agilent-AVO-BlackStar-Brymen-Chauvin Arnoux-Fluke-GenRad-Hameg-HP-Keithley-IsoTech-Mastech-Megger-Metrix-Micronta-Racal-RFL-Siglent-Solartron-Tektronix-Thurlby-Time Electronics-TTi-UniT
 

Offline DavidKo

  • Frequent Contributor
  • **
  • Posts: 295
  • Country: cz
Re: Agilent 34401A 6.5 Digit MM Problem
« Reply #4 on: June 06, 2023, 11:38:41 am »
Look for the manual on the Keysight web site. There is also quick start guide.

The navigation in the menu is little bit non-intuitive. It took me a while till I get familiar with it.
 