| babysitter:
Call me interested in switching off the current source! |
| alex.forencich:
Interesting. I installed the IntuiLink package in a VM and found the example files. Looks like it uses 4 different POKEs: DIAG:POKE -2,4281,0 with no particular comment, presuming 'close front end relay' DIAG:POKE -2,4281,4 with comment 'open front end relay' DIAG:POKE -2,3512,0 with comment 'force front end conf to change' DIAG:POKE 1,0,0 with comment 'force front end conf to be sent' Basically, the code works like this: put instrument in 2 wire ohms mode and send first poke put instrument in 4 wire ohms mode wait take current source on measurement send other 3 pokes to open relay wait take current source off measurement subtract |
| babysitter:
Unrelated rant: Just called Agilent Service, I sent them a workplace 34401A for calibration on Feb 9th, ordered a calibration in their InfoLine service upfront, had a written order in the package and thought everything is well... Wrong. They send me a email with a quote asking for confirmation, which I answered on 26th. Still not enough, today i got: Received 02/12/2014. Awaiting quote approval Called, they skipped my mail but promised everything goes fine from today. :wtf: |
| coromonadalix:
Just found an vb example of this from an intuilink installation file IMPORTANT NOTE: This program describes an UNSUPPORTED CAPABILITY of the Agilent 34401 Multimeter. Agilent does not warrant or support the performance or use of this feature in any manner. Some customers want to measure lower ACI than the 1 A range allows. The lowest specified measurement on the 1 A range is 1% of range, or 10 mA (see footnote 4 in data sheet). The code below enables a hidden 10 mA ACI range. ------------------------------------------------------------------------ 10 ! 20 ! 30 ! The following program makes several ACI readings using 40 ! the undocumented 10 mA range. 50 ! 60 ! The current range is enabled using the BUS and a DIAG:POKE 25,0,1 70 ! command. Once enabled, the range can be accessed over the 80 ! BUS or from the front panel. You can still access the range 90 ! even if power is cycled. To disable the range you need to 100 ! to send a DIAG:POKE 25,0,0 command. 110 ! 120 ! BCG 1/6/93 130 ! 140 Nrdgs=10 ! NUMBER OF RDGS 150 ! 160 ALLOCATE Rdgs(1:Nrdgs) 170 ! 180 ASSIGN @Dvm TO 723 190 ! 200 ! DIG OUT PREVIOUS ERRORS 210 ! 220 CLEAR 7 230 REPEAT 240 OUTPUT @Dvm;"SYST:ERR?" 250 ENTER @Dvm;A,A$ 260 PRINT A,A$ 270 UNTIL A=0 280 PRINT "-----------------------------------------------------------" 290 ! 300 ! SET UP 34401A 310 ! 320 CLEAR @Dvm 330 OUTPUT @Dvm;"*RST" 340 WAIT .5 350 ! 360 ! 370 OUTPUT @Dvm;"DIAG:POKE 25,0,1" ! Allows access to 10 mA 380 OUTPUT @Dvm;"CONF:CURR:AC .01,MIN" ! ACI from the bus and 390 OUTPUT @Dvm;"TRIG:COUN";Nrdgs ! front panel. 400 OUTPUT @Dvm;"READ?" 410 ! 420 FOR I=1 TO Nrdgs 430 ENTER @Dvm USING "#,K";Rdgs(I) 440 PRINT I,Rdgs(I) 450 DISP I,Rdgs(I) 460 NEXT I 470 ! OUTPUT @Dvm;"DIAG:POKE 25,0,0" ! DISABLE 10 mA RANGE 480 PRINT "-----------------------------------------------------------" 490 END |
| Brandy:
What is the advantage of the vb program compared to the poke mentioned from macboy in the first post? Btw, the "autorange" function is working upwards only. e.g. if you measure 10mA 1kHz and input a higher current value, the 34401 will change to the 1A range automatically (if you're in autorange mode). If you change from lets say 15mA/1kHz to 5mA/1kHz it will NOT switch to the 10mA range! This has to be done manually. |