I recently bought an Agilent 34401a which shows some strange behavior. Sometimes if I do VAC measurements it changes unexpectedly to Resistance Measurement.
It passes the selftest without errors.
Does anyone know the reason for this problem?
Thanks for your help
Nicolas
-
The two buttons are next to each other ACV (SW613), 2W ohms (SW614) but different MCU scan lines.
So I would suspect the ohms pushbutton sticking/shorting or a solder bridge on the front panel PCB. Try gently pushing and tapping to see if the ohms button is shorting.
I take these apart and clean up the front panels, toss the keypad in dishsoap and water to clean it up.
Sometimes I find something was spilt inside, or loose strand of copper wire got in.
-
Thank you for the advice floobydust.
I removed the Frontpanel and cleaned everything, but the problem occured again. While measuring DCV, it also switched to 2W ohms after some time.
-
Leave the front panel disconnected (Remove the cable ) and try to control the instrument over GPIB or RS232 and see if it still switches to 2W Ohm.
-
I tried what you suggested HighVoltage, but It still switches to 2W Ohm after some time with disconnected ribbon cable and control via RS232.