Products > Test Equipment

Agilent 34410A Noisy fan

<< < (5/6) > >>

RolandK:
The 34410A / 34411A seem to have 30 / 10 mm fans.
The 34465A / 34470a seem to have 40 / 10 mm fans. Both just measured on the backside.
The E3449a seems to have 80 / ? mm fans.
In the https://www.ebmpapst.com/content/dam/ebm-papst/media/catalogs/products/Catalog_Compactfans_US_en.pdf is only a 40 / 10 mm fan, no 30 /10. So no luck with this. :(

HighVoltage:

--- Quote from: Odd on April 16, 2023, 06:23:23 pm ---can you please tell me the papst fan model you have used for 34410A ? (I'd like to avoid disassembly just to figure out the dimensions and voltage & wattage/current)

--- End quote ---
It has been too long since I did this mod.
I just looked at my notes, this was not a Papst fan.
I used Papst fans for other Agilent gear.

The fan you need for the 34410A and 34411A is a 12V 35x35x10mm.

The model that I installed was a:
Sunon MF35101V1-1000U-A99 Axial fan 12 V DC 12.24 m³/h (L x W x H) 35 x 35 x 10 mm

RolandK:
In my 34411a with serial number starting MY48 is a 1204KL-04W-B40 12V DC 0.1A from NMB-MAT (Minebea-Matsushita) which is 30x30x10mm.

Did they change the size, e.g. in later versions? Or did you modify the fan-mounting with a drill and dremel? :-BROKE

HighVoltage:
Nothing was changed, definitely not with a dremel.  :-DD

I am traveling right now. When I am back, I will open one instrument and take some pictures.

RolandK:
You had the same fan, here are your pictures: https://www.eevblog.com/forum/repair/agilent-34410a-repair-attempt-of-2-dead-meters/msg820491/#msg820491

i ordered now sunon HA30101V3-1000U-A99

Navigation

[0] Message Index

[#] Next page

[*] Previous page

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod