EEVblog Electronics Community Forum

Products => Test Equipment => Topic started by: yo0 on May 14, 2019, 05:44:10 pm

Title: Agilent 54624a any good?
Post by: yo0 on May 14, 2019, 05:44:10 pm
i have the opportunity to buy an Oscilloscope Agilent 54624a in near mint condition. any opinions about this model? it is better than a tektronix 465?


any comments will be really appreciated.

best regards.

Pio
Title: Re: Agilent 54624a any good?
Post by: gslick on May 15, 2019, 01:26:04 am
Dave did a video about the Agilent 54622D five years ago, back in March 2014

EEVblog #591 - Agilent 54622D Retro Mixed Signal Osciloscope Review & Teardown
https://www.youtube.com/watch?v=aVNfFewFn_Y (https://www.youtube.com/watch?v=aVNfFewFn_Y)

The difference between the 54622D that is the subject of that video and the 54624A you are asking about is that the 54622D has 2 analog channels plus 16 digital channels, while the 54624A has 4 analog channels and no digital channels. Both are 100 MHz, 200 MSa/s, 4MB sample length scopes.

Agilent Technologies 54600-Series Oscilloscopes Data Sheet
https://literature.cdn.keysight.com/litweb/pdf/5968-8152EN.pdf (https://literature.cdn.keysight.com/litweb/pdf/5968-8152EN.pdf)

If the price is right a 54624A could still be a perfectly usable scope, depending on what you need.
Title: Re: Agilent 54624a any good?
Post by: yo0 on May 15, 2019, 02:41:23 am
Dave did a video about the Agilent 54622D five years ago, back in March 2014

EEVblog #591 - Agilent 54622D Retro Mixed Signal Osciloscope Review & Teardown
https://www.youtube.com/watch?v=aVNfFewFn_Y (https://www.youtube.com/watch?v=aVNfFewFn_Y)

The difference between the 54622D that is the subject of that video and the 54624A you are asking about is that the 54622D has 2 analog channels plus 16 digital channels, while the 54624A has 4 analog channels and no digital channels. Both are 100 MHz, 200 MSa/s, 4MB sample length scopes.

Agilent Technologies 54600-Series Oscilloscopes Data Sheet
https://literature.cdn.keysight.com/litweb/pdf/5968-8152EN.pdf (https://literature.cdn.keysight.com/litweb/pdf/5968-8152EN.pdf)

If the price is right a 54624A could still be a perfectly usable scope, depending on what you need.

thank you, the asking price is approx 150 usd, the main use will be for analog and audio stuff. i wonder if the tek 465 is a better equipment despite the cursors and readouts in the agilent.

best regards

Pio
Title: Re: Agilent 54624a any good?
Post by: siggi on May 15, 2019, 10:24:58 pm
As a first or only scope, I'd go for the 54624a every day of the week. IMHO analog scopes are more pleasant to use. There are however things you can do with a digital scope that an analog scope just won't do, such as looking at slow or one-time events and looking at pre-trigger signal.
The 54624a also will do automated measurements for you, it has much more capable triggering than the 465 and it can do FFT - it's not super-awesome, but it does it.

I have a 54622D, and while my Tek 2467 is my favorite scope in usability, I can say I don't hate the usability of the 54622D. It drives pretty much like an analog scope...
Title: Re: Agilent 54624a any good?
Post by: yo0 on May 15, 2019, 11:30:42 pm
As a first or only scope, I'd go for the 54624a every day of the week. IMHO analog scopes are more pleasant to use. There are however things you can do with a digital scope that an analog scope just won't do, such as looking at slow or one-time events and looking at pre-trigger signal.
The 54624a also will do automated measurements for you, it has much more capable triggering than the 465 and it can do FFT - it's not super-awesome, but it does it.

I have a 54622D, and while my Tek 2467 is my favorite scope in usability, I can say I don't hate the usability of the 54622D. It drives pretty much like an analog scope...

thank you very much, i will buy it  ;D

best regards

Pio