Tony_G:
You can just change the drive letter used by MSCDEX in autoexec.bat - "/L:F" will mount the drive to F: - You might be able to use "assign" to redirect F: to the cdrom drive but it really is just easier to change the letter in autoexec.

I would simply write the Win98SE boot disk image to a floppy, change the letter to F, boot, and run cdromupg - You don't need to change anything on the device, just the boot floppy. You can then reboot and go back to whatever config you want.

TonyG


BoscoLab:
I finally finished the update on the HP 54845A, now it has A.04.50. Thanks for the help!  :)

Rudolf

Tony_G:
Good to hear - Glad I could help.

TonyG

ARF:
I'll update the topic. I recently received 54845A. At first glance, everything works, but during calibration I received an error on one channel. Any advices? I remember that in 54820 oscilloscopes it was necessary to clean the sockets for ADC modules.

nctnico:
You can try and clean the sockets but it is also possible the input module is defective.