Products > Test Equipment
Agilent 6813B, Agilent 6573A broken by insulation test.
(1/2) > >>
HackedFridgeMagnet:
Hi
Some sparkys just did some rewiring work in our lab and I am pretty sure they have damaged our PSUs when they did their insulation testing.

I have 2 x Agilent 6573B completely dead. (Hopefully a fuse inside). Anyone got a schematic?

1 x Agilent 6813B that constantly reboots. Anyone got a schematic?

They were all working last time I looked a few months ago, then the sparkies started work, a few weeks later the sparkies said one unit had stopped working.
The next day 2 more units had stopped working.

Luckily the biggest psu  the 3 phase version of the 6813 is still fine.



ogden:
First question here is: are they sparked equipment correctly? Ask them which standard(s) they tested against. Shall be UL 3111-1 (electrical equipment intended for professional, industrial process) or  IEC 1010 or EN 61010. I am afraid that damage is far worse than just simply fuse. No, I don't have schematic and doubt any1 besides Keysight have. You shall start contacting their service or sales

Zucca:
Were the devices on while the "sparkys" were playing?
HackedFridgeMagnet:
I doubt they sparked (tested) the installation correctly as they damaged my power supplies.
I am almost but not quite certain it is their fault. They did a good job otherwise and I want to stay on the right side of them. As they are nice and do a lot of work for us. Also two of them were quite young.

The standard they test to is a local standard, named AS/NZS3000. Which probably has derived the insulation test based on some international standard.

The test is pretty easy.
Just Apply 500vdc (high impedance source) between each of the 'live' conductors(including Neutral) and Earth and make sure there is at least 1 Meg ohm of resistance.

In Australia we have a slightly modified TN−C−S Earthing system. So at one point in the building the Earth is shorted to the Neutral.
If the link between Earth and Neutral isn't disconnected before testing and your device in in circuit then you will get 500v into the device from Active to Neutral.
I have myself once damaged an oven doing this test, so these days I am careful about it. Most devices wouldn't have a problem with this.
One thing I forgot to mention was that the devices are hard wired in, and have circuit breakers in, the AC devices have RCDs on the output.
They all have power into the box.

I was hoping they were galvanically isolated but now I am not so sure.
I had a quick look for schematics but haven't found any yet.


tombi:
So you don't mean 6573A? There is a service manual with schematic on the Keysight page.
https://www.keysight.com/en/pd-839233-pn-6573A/2000-watt-power-supply-35v-60a?pm=PL&nid=-32611.384516&cc=AU&lc=eng

Ditto the 6813B
https://www.keysight.com/en/pdx-2755769-pn-6813B/performance-ac-power-source-1750-va-300-v-13-a?pm=rsc&nid=-32114.1216486&cc=AU&lc=eng