TomT:
Hi Ordinaryman,
I used to repair the RF boards on the ESG series to component level, and have lots of experience with the reference boards. The only repairs to the reference boards that were ever needed was to replace the Reference Oscillator, the 1GHz oscillator transistor, one of the VCO voltage drive transistors (only used on some boards), Ext. Ref. Input transistors, or diodes. I personally own an E4432B ESG-D, and a recently acquired E4437B ESG-DP. There was a problem with unlevel on the E4432B, but with the knowledge of repairs of output boards I did at work, I purchased a few of the HP proprietary RF amps before they became unavailable.
First, knowing the full model number, and options would be very helpful in diagnosing this problem. If you can supply the Reference Board part number, even better.
The 1GHz VCO, is located on the Reference board, at the opposite edge of the board from the connectors It is easily identifiable by the hard line coax soldered to most boards, and is the fixed LC for the oscillator. Others have a small VCO module. The 1GHz is divided down to 10MHz, compared to the divided down, frequency and phase compared to the 10MHz Reference Oscillator, and low pass filtered to create an error voltage, which is then applied to the voltage control of the 1GHz VCO. If the Reference Oscillator is too far out of tolerance, the 1GHz oscillator is unable to lock. Likewise, if the 1GHz free run frequency is too far off, it won't lock.
The YO loop may lock once the 1GHz is locked. It’s just best to start with the reference board first, as it is the source for the entire generator.
As for the Ext. Ref. not locking the 1GHz VCO, the best place to start is at the Ext. Ref. input to the board. The input circuitry may have been compromised from an improper input supplied, which I have repaired in the past. It’s also possible that phase detector circuit has issues, but I did not experience any problems in this area when I was working on them.
As for the display, play with the contrast controls. That was one problem that messed with me and others that I worked with, especially after replacing the display. Also, if I recall, there are two different part numbers for the displays, and I don’t recall if they are interchangeable. I believe that one of the part numbers was used in newer units as the older display became unavailable, or a better or cheaper display was found.
If you plan to attempt RF board repairs yourself, when reassembling the clam shell, be careful about over-tightening the screws, as they are stainless steel screwing into aluminum. Start by starting all of the screws before tightening, and be careful not to cross-thread them. Then start tightening from the center screws, and work your way out to the edges.
grendle373:
Great info!
Question, I have an e4432A, I repaired a display issue ( Bad LCD, Blown Q601, Q602), Display is fine, everything else works correctly and is in spec. , except I get a "Fail 601 power search" when running the IQ mod cal.
I replaced the output board (usually the culprit) with a known good one ( passes IQ cal in a similiar unit), and still get the failure, Also replaced the Refrence board with a known good one, no change. Unit has option UND (BB generator). Output level drops about 11 db when the IQ mod is switched on. Scratching my head on this one......
mbielman:
I know this is an old post but are these from the CAL software package?
SERV:PRODUCTION:CAL:BEGIN;
SERV:PRODUCTION:CAL 165,2,4000000000.000000;
SERV:PRODUCTION:CAL:END;
SERV:PRODUCTION:CAL:STORE 165;
SERV:PRODUCTION:CAL:PUP;
They are not in the SCPI docs.
Thanks!
vaualbus:
So basically this commands invalidate the cal data If I got that part correct from previous comments? I too recently purchase a 2Ghz version and maybe I wanna "upgrade" to 3 or 4ghz. Also are there any way to dump the ROMs from the serial/GPIB interface? It's also strange that we don't have a debug serial port somewhere in the intruments!
Speaking about hacking, as any one ever tried to hack the software options to add them to the instrument?
mbielman:
What I'm trying to ask is 'Where do these commands come from? How do you know about them? Are they documented?'
Maybe they are from the software .log files? Or the test window itself?
Assume there is NO docs available for these.
(what does the PUP command do???)