| Products > Test Equipment |
| Agilent E7495 linux root account |
| << < (85/91) > >> |
| nsd_c:
Hi all, I have an E7495A that I purchased a few years ago off of eBay used that I've used successfully for one- and two-port measurements, distance to fault, and as a general spectrum analyzer. It has come in very handy for all of these functions. I decided recently to extend its functionality by purchasing a power probe, the Agilent 8481H (-10 to +35dBm), from eBay. The seller claimed that it was working, and seems reputable, however, after running the 'zero' and 'calibrate' functions, this power sensor reads around -19dBm with the power reference turned on (and the sensor hooked directly to the power reference output). From reading the manual, I believe it should read 0dBm (1mW @ 50 Hz is the rated output). It does read very low with the power reference turned off (about -40dBm). For lack of any other useful instrumentation, I tried measuring the power output using an RTL-SDR and about 39dB of attenuators chained together. The RTL-SDR showed -28.94dB of input power (appearing as a peak at around 50 MHz), which gives +10.09dBm coming out of the '7495A... or 10mW instead of 1mW (a ten-fold increase). I did do a continuity check of the probe cable, and it seems OK. For lack of instrumentation to check with (I do not have another power meter or probe), I'd like to check the power sensor reading using not the power reference port, but the E7495A's built in generator (if possible). I am able to telnet into the instrument (thanks to info on this thread). I have seen some of the Java GUI reversing bits on this thread as well, so I was curious to try to find out if there is a back-door way of commanding the power reference on while the spectrum analyzer is running, or likewise turning the generator on when the power meter application is running. To that end, I found this bit in PowerMeterScreen.class: --- Code: --- MenuItem createPowerRefButton() { return new MultiStateActuatorButton(PowerMeterMeasurementSettings.instance().getPmRef(), getContextString("powerMeterCalibrate.powerRef")); } --- End code --- If I dig into MultiStateActuatorButton, I can see that the first argument to the constructor is a 'ListActuator' that the MultiStateActuatorButton uses to "do the thing" when the button is pressed: --- Code: --- public MultiStateActuatorButton(ListActuator paramListActuator, String paramString) { super(paramListActuator.getLabel(), paramString); this._actuator = paramListActuator; addActionListener(this); } public void actionPerformed(ActionEvent paramActionEvent) { if (!isFocused()) { getMenuPanel().focusButton(this); } this._actuator.increment(); <<< The magic repaint(); } --- End code --- Skipping some in-between stuff, the getPmRef() method returns this: --- Code: ---private ListActuator _pmRef = new ListActuator("pwrMeter", "pmRef", Text.SCREEN_POWER_POWER_REF, new Value[] { Value.createValue(Text.ON, 1), Value.createValue(Text.OFF, 0) }); --- End code --- where the ListActuator constructor takes these args: --- Code: ---String paramString1, String paramString2, String paramString3, Value[] paramArrayOfValue --- End code --- I got kind of lost reading through ListActuator, but it looks like ListActuator ultimately sends a command to the elgato server with the send() method of the WebPlugConversion() class. It looks like WebPlugConversion then uses the Command() class to build a string that it sends: --- Code: --- public static Command makeSetCommand(String paramString) { Command localCommand = new Command("set"); if (!paramString.equals("global")) { localCommand.addProperty("topic", paramString); } return localCommand; } --- End code --- I likewise got a little lost trying to figure out what the final command string ends up looking like. However, after digging through this thread some more, I came across reply #126, where user DogP shared some observations: --- Quote ---I also took a look at the remote GUI... [... snip ...] I was interested in how it communicated, so I sniffed the packets. It looks like most stuff goes over port 5028. The commands and acknowledgements go over TCP, and the spectrum display data goes over UDP. It looks like there’s some status (power, GPS, etc) coming out 1027 as well. [... snip ...] The commands going over 5028 don't look to be any standard I'm familiar with. Some examples are (without the quotes, <\n> is hex 0A): “cmd: local<\n><\n>”, "cmd: set<\n>topic: display.global<\n>printFilter: 0<\n><\n>", etc. If you dig through the elgato binary, you can see a bunch of available commands (and you can test them out using nc). --- End quote --- Armed with this information (namely, what port to look at), I found that if you telnet into port 5028, pressing buttons on the front panel has the effect of the system echoing on that port these commands. So, if I turn the power meter on through the front panel, I see: --- Code: ---cmd: set topic: pwrMeter pmRef: 0 --- End code --- get printed to the terminal. I am able to make the ‘get’ commands work: if I type (for example) --- Code: ---cmd: get topic: spectrum cf: --- End code --- the instrument responds with: --- Code: ---cmd: set topic: spectrum cf: 1955000 --- End code --- However, I can’t seem to get ‘set’ commands to work—attempting to change any parameter causes the instrument to do… nothing. The only command other than ‘get’ that works is ‘cmd: closeGui’, which causes the connection to be closed by the host (instrument). I’ll note that when I first connect to this port, the instrument prints: --- Code: ---cmd: remote cmd: setActive meas: cmd: notifyEvent event: startupComplete --- End code --- I was hoping to use Wireshark to capture commands sent by the Java GUI running on my PC, but I can’t seem to get the GUI to work quite 100%. It starts (in a very old Debian Squeeze VM running in a Windows host by way of VirtualBox), but hangs at “Please wait while the measurement is loading…”. It does work insofar as the GUI running on my PC does show the correct instrument mode (i.e., main menu, spectrum analyzer, power meter, etc.). User DogP’s aforementioned post did mention --- Quote ---[…] the remote GUI […] didn't seem to work on Windows with Java 8 (even using the commands further up in the thread). Rather than dig into it, I installed Java 1.4.2 in a VM, grabbed the jre directory, and explicitly ran RemoteGui.jar using that version of java ("jre\bin\java -jar RemoteGui.jar"). [...] It looks like most stuff goes over port 5028. The commands and acknowledgements go over TCP, and the spectrum display data goes over UDP. […] I had to make sure Windows allowed the data through the firewall (without it, the TCP commands would go through, but the UDP data wouldn’t, so the GUI would never show the spectrum). Also, Windows 7 temporarily goes back to basic graphics (disabling Aero) when you run Java 1.4.2. --- End quote --- I tried installing JRE 1.4.2_20, but invoking the application with the command line given earlier in this thread failed with --- Code: ---Exception in thread "main" java.lang.NoClassDefFoundError: elgato/gui/MainWindow --- End code --- I found that this works instead, though: --- Code: ---java.exe -Dserver=192.168.1.186 -DlocalAddress=192.168.1.1 -classpath "./*" elgato.gui.MainWindow --- End code --- However, clicking around in the Java GUI doesn't do anything. I did try copying over the RemoteGUI.jar file, but this causes its own problems. I am able to change the spectrum analyzer center frequency by pressing the 'up' and 'down' arrows on my computer, and I can sniff out the 'cmd: set' command that are being sent when I do this in Wireshark. I'll note also that my instrument has a significantly newer version of SW than what's available for download from Keysight's site (A.06.00, from 2006, instead of A.01.60, from 2003). Anyway, I thought I'd share that progress, and ask a few questions: * Does anyone know why the power ref output might be generating 10mW instead of 1mW? * Does anyone already know of a back-door way of doing what I'm trying to accomplish? * Any ideas why the instrument doesn't like my 'cmd: set' commands? |
| kirill_ka:
Hi, lets proceed with one problem at a time :) I suspect RTL-SDR is not a reliable tool to do power measurements. Try enabling power ref as I described here: https://www.eevblog.com/forum/testgear/agilent-e7495-linux-root-account/msg1169477/#msg1169477 Then measure the reference using E7495 itself (don't forget to add some attenuators!). |
| kirill_ka:
--- Quote from: nsd_c on August 11, 2021, 05:58:14 pm --- functions, this power sensor reads around -19dBm with the power reference turned on (and the sensor hooked directly to the power reference output). From reading the manual, I believe it should read 0dBm (1mW @ 50 Hz is the rated output). It does read very low with the power reference turned off (about -40dBm). --- End quote --- My understanding is that HP8481H = HP8481A + 20dB attenuator. So it could be a bug with handling this particular type of sensor. Though, I don't know where -19dBm comes from. Maybe the calibration fails? Or maybe HP8481H is supposed to have 19dB attenuator? What's the actual reading? Is it exact -19.00 dBm? |
| Ick:
Pieroc91, From the logs it seems you have a problem with a corrupt file system ( the lose serial cable you have fixed already); all the jffs2 error messages. You need to reinstall the firmware and recreate the filesystem. This is done the following way: - insert the firmware to preferably PCMCIA slot(CF card in PCMCIA adapter). On my unit the upgrade from CF slot would not work properly. - start/ reboot the unit - on the upgrade screen with the continue yes button, press the LOWER RIGHT soft button next to the LCD instead. Yes, there is no marking, description or hint on the upgrade screen that this button has a function. Press this lower right soft key and the complete flash will be erased, file system re-created and firmware files newly written. It seems a backup of the license file is automatically been made, I would save it in advance anyway. I had the similar issues with my unit - file system errors with same message. That caused all kind of odd behaviour, e.g could not properly access CF slot and FS anymore , no dhcp, no upgrade possible (needed to go via dip switches on the COU board) After the new filesystem created, all was hunky dory. Error messages were gone during boot and all properly working. BR ick |
| pieroc91:
Hi everyone! For anyone fighting with the Remote GUI i've made a package with two executables to easily connect to the device. They are just batchs converted to exe The automatic local IP code is: --- Code: ---color 17 echo off for /f "tokens=1-2 delims=:" %%a in ('ipconfig^|find "IPv4"') do set ip=%%b set ip=%ip:~1% cls set /p AgilentIP=E7495 IP Address: rem echo %AgilentIP% echo Local IP Address: %ip% (Autodetected) echo. echo Press any key to connect pause>nul jre\1_8_0\bin\java.exe -Dserver=%AgilentIP% -DlocalAddress=%ip% -classpath "./egclient" elgato.gui.MainWindow --- End code --- And the manual Local IP is: --- Code: ---color 17 echo off cls set /p AgilentIP=E7495 IP Address: rem echo %AgilentIP% set /p ip=Local IP Address: echo. echo Press any key to connect pause>nul jre\1_8_0\bin\java.exe -Dserver=%AgilentIP% -DlocalAddress=%ip% -classpath "./egclient" elgato.gui.MainWindow --- End code --- It's also bundled a version of JRE 1.8.0 that works with Windows 7 and 10 without a problem. It's available to download from here: https://drive.google.com/file/d/1MNUHC3evNMRjFCOC_0AetulkUSO1vfGo/view?usp=sharing |
| Navigation |
| Message Index |
| Next page |
| Previous page |