joeqsmith:
After selecting save.   I never tried to restore.
Forzaman:
Ok. I thought that is how you got to the second EEPROM save and restore menu where you can select the location to save the backup file. I don't get that menu... maybe I have older firmware. By default it saves the backup to D:/calfiles, and the name of my file is eebudat.dat not EEPROM.dat.
joeqsmith:
Did you select the Specify Location like I show?  I wanted to save mine to my local PC, not the PNA and that may be the reason for the difference.  You could also check that version is the same. 

Note how that second EEPROM save is pointing to E:\MXCAL which is the my desktop PC.   
joeqsmith:
Also, if you don't mind posting the two files (before and after).  If you add the extended range and realign it again, I would like to have that file as well. 

If Keysight has some sort of software for that power meter, it would be very nice to see how tight your meter is compared with the VNA.   The problem is you would need to use CW and manually set the PNA.  More trouble than it's worth.   
Forzaman:
I don't get the menu that asks where to save the EEPROM data. I get the first menu (screen shot attached). When I select Backup EEPROM a message pops-up saying the EEPROM was backed-up (no crash), but didn't say where. When I selected Restore EE from backup, a pop-up window appeared asking for the password, after that it asked if I wanted to restore the  EEPROM from D:\calfiles\eebudat.dat without providing an option to select from another location. I do not get the 2nd menu you show where you can select the file location. It seems my EEBackup program defaults to the D:\calfiles folder.