Author Topic: AGILENT E9300B and E4418B problem.  (Read 238 times)

0 Members and 1 Guest are viewing this topic.

Offline MikeP

  • Regular Contributor
  • *
  • Posts: 119
  • Country: ua
AGILENT E9300B and E4418B problem.
« on: March 04, 2020, 05:33:32 pm »
 Hello all.

 Two problems in one place.

 E9300B sensor with 30dB attenuator. Usually gives readings ten times more than the actual value. But  VERY rarely do I get the correct value. The sensor is disassembled, unfortunately there is nothing obvious. We have no experience with this device and I want to be careful. Perhaps the sensor is not hopeless. So I ask for help.

 And the second problem. The power meter often does not respond to the power button. All tests passed. The keyboard, connector, and ribbon cable are checked. Now we are checking the main board.

 I would be grateful for any help.
 

Offline MikeP

  • Regular Contributor
  • *
  • Posts: 119
  • Country: ua
Re: AGILENT E9300B and E4418B problem.
« Reply #1 on: March 08, 2020, 07:12:25 am »
 A small update. The problem with the power button has been resolved. One of the divider resistors was defective. Therefore, the logic gate received insufficient voltage. Everything is simple.

 Unfortunately, there is no solution with the sensor. But there is hope. Deformation may result in obtaining correct data. We still hope for community help. Thanks.
 

Offline kirill_ka

  • Regular Contributor
  • *
  • Posts: 228
  • Country: ru
Re: AGILENT E9300B and E4418B problem.
« Reply #2 on: March 08, 2020, 06:04:32 pm »
Quote

Usually gives readings ten times more than the actual value.

Please, describe your experiments and setup in detail. Do you get different readings while measuring the 50MHz calibration?  How do you know what's the actual power? Do you have another power sensor to compare?
« Last Edit: March 08, 2020, 06:06:07 pm by kirill_ka »
 

Offline MikeP

  • Regular Contributor
  • *
  • Posts: 119
  • Country: ua
Re: AGILENT E9300B and E4418B problem.
« Reply #3 on: March 08, 2020, 07:25:31 pm »
 Yes, we have a second power meter and two GHz generators.

 At the moment, 4418 without any problems.

 Most likely, the problem is in the sensor. By deforming the sensors PCB, we get true results. Under the microscope, everything is perfect.

 But until I hear experienced comrades, I don’t want to do anything. The price of the error is very high.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf