Minsoon:
Thank you very much.
I tried booting Win XP... recovered it with Recovery... and installed SW 5.71 on it again.
It works very normally. I'll have to do the calibration tomorrow.

And eevblog said that you can save the option by doing the xdelta patch.

xdelta3-3.0.11-x86_64.exe -d -s agScope.exe delta agscopeNew.exe

I ran it like this, overwritten the changed New file, and ran it...
Option only comes out as 0.

Please tell me how to save this option ^^

ARF:
See your PM.

-------------

If you have the 1 generation of the ACQ board = you can enable only 8Mpts per channel, if the 2 gen = 64Mpts.
The 1 gen has TQFP DRAM ICs, the 2 gen has more modern TSSOP DRAM ICs. You can share a photo of your ACQ.

You can enable 1G bandwidth via disordering 1 resistor. I'm sure, you already found this information.

Minsoon:
Thank you very much.
Thanks to you, all options were installed.
The Acq board seems to be the 1st generation.

And the self-test was OK, but the trigger part of the calibration all failed.
I used a BNC coaxial cable when calibrating. Will this be a problem?

ARF:
Did you have this problem with original Windows 98? It should work fine with Win XP, but... Yes, this is the 1-gen ACQ with 8Mpts/channel.

You still can unlock the 1Ghz BW, but first you need resolve your trigger problem.

Minsoon:
Ok..Thanks..

I will solve the trigger problem first and then post it.