Products => Test Equipment => Topic started by: Palmsquirrell on April 08, 2014, 11:51:45 am

Title: Agilent / HP CDMA base station E6380A ; replacing memory batteries.
Post by: Palmsquirrell on April 08, 2014, 11:51:45 am
Agilent/HP  CDMA Base station E6380A

I get the following : " AA memory backup batteries are low "

I do not have a owner's manual, but I did look at an online manual but have not found the instructions on how to replace the batteries.

Thanks
Title: Re: Agilent / HP CDMA base station E6380A ; replacing memory batteries.
Post by: Towger on April 08, 2014, 12:12:02 pm
You need to improve your google skills.
From page 31 of the manual: http://www.home.agilent.com/upload/cmc_upload/All/6C06UGE6380-90016.pdf (http://www.home.agilent.com/upload/cmc_upload/All/6C06UGE6380-90016.pdf)
 
Do not use rechargeable batteries.
To change the AA batteries, use the following procedure:
1. Turn off power and unplug the Test Set.
2. Remove the six screws in the rear panel using a TX-15 TORX (R)
screwdriver.
3. Remove the rear cover.
4. Replace the AA batteries.
5. Replace and reconnect the rear panel. Dispose of used batteries
properly.
Title: Re: Agilent / HP CDMA base station E6380A ; replacing memory batteries.
Post by: Palmsquirrell on April 08, 2014, 04:20:19 pm


I looked at that same manual and saw batteries mentioned in the Index as being on page 31. So I went to page 31 on the pdf instead of page 31 on the actual manual.

I normally check the page number printed on the paper as it does not correspond to the page # on the PDF but this time I forgot all about it. |O

Thanks.