Hi!
If memory serves me, in a PDF article Agisight mentioned this about gaussian vs maximally flat (or something like that) frequency response characteristics for digital scopes. Less than 1GHz bandwith often being gaussian with -3dB at typical specified bandwith. I also think I read about the use of DSP to straighten out the response that is displayed in some products (could have been some other brand though).
I have a DSOX3032 wich is supposed to have 350MHz bandwith but when looking in the calibration certificate document from the factory, that spec says -0.27dB and -0.32dB for channel 1 and 2 down at 350MHz (200mV/div). The document mentioned -3dB as lower limit for that particular test.
I also looked at Daves video where he plays with the Jim Williams pulse generator, indicating that his 500MHz Agisight scope has a -3dB bandwith of 600-700MHz or so.
So, is this the norm or just random "overspecification"?
Maybe I actually have a 500MHz scope according to the usual way of specify?
Or did they forget to lock my scope to 350MHz.. the scope can be software upgraded to 500MHz (license key).
Happy measuring!