Today, I'm looking at the temperature measurement capabilities of the Aim-TTi 1908. The unit can be used with
PT100 thermistors in 2-wire or 4-wire configuration, or with
PT1000 thermistors in 2-wire or 4-wire configuration. It can measure from -50C to 400C with either thermistor. The accuracy is 0.05% +- 0.5C:
Ideally, I wanted to get a 4-wire PT1000 thermistor for the highest accuracy, but those are difficult to find at an affordable price. After some research, I found out that 4-wires aren't as important with PT1000 due to the higher resolution it provides, so I went with a PT1000.
I picked up an
Atlas Scientific PT1000 temp probe (
datasheet), plus a BNC to banana adapter.
As far as accuracy goes, how is that calculated? Does the error from the DMM (0.05% +- 0.5C) add linearly with the thermistor accuracy (0.15%)?
Before measuring, I first went into the utilities menu and changed the temp reading mode to 2-wire (defaults to 4-wire).
Then you hit temp option and hit right to go into PT1000 mode. Here is the temperature reading at room temp, seems "about right":
In Farenheit:
In Celcius:
As far as high temperature measurements, my plan was to test the probe against my TS100 soldering tip set to 200-400C.
Unfortunately, even with the fattest tip for the largest surface area, and me pushing down hard as I can, the temp was rising very slowly (0.1C per 1 sec). After quite some time, it settled and wasn't able to measure higher than 100-120C (with iron at 400C):
I know for a fact that the temp reading on the TS100 soldering iron is fairly accurate by using it over the years. It also melts solder like butter. Perhaps either the DMM is faulty, or the thermistor is faulty, or I'm just doing something wrong.
As as sanity check, I switched to 2-wire resistance mode while measuring temp, and the resistance values were correct (e.g. ~1250 ohms at 64C).
Temp calculation equation: